Abdominal
Diskussionsliste
-
(2026) Chen | Phase-specific reconstruction optimizes low-dose liver VPCT on PCCT
In 62 patients with liver disease, PCCT-based liver volume perfusion CT (VPCT) enabled a one-stop workflow combining perfusion with morphologic assessment. Compared with a simulated conventional four-phase CT plus perfusion workflow, effective dose decreased from 31.9 ± 8.5 to 25.9 ± 6.7 mSv, an approximately 18.8%…
-
(2026) Herold | UHR PCCT reconstruction settings for rectal cancer imaging
In this prospective pilot study of 19 patients, UHR PCCT image quality in rectal cancer was strongly influenced by both slice thickness and monoenergetic level. Thicker slices increased SNR and reduced noise, while lower keV increased tumor-to-muscle CNR at the expense of higher noise; CNR ranged from 4.85 ± 2.43 at 3…
-
(2026) Brandt | Reduced beam hardening in urogenital imaging with photon-counting CT
In 35 patients undergoing both PCCT and conventional EID-CT, four radiologists evaluated image quality and beam-hardening artifacts in the lower pelvis and perirenal region. PCCT significantly improved subjective image quality: median scores increased from 3 to 5 in both the perirenal region and lower pelvis (P < .001).…
-
(2025) Braaksma | Pediatric abdomen PCCT protocol development
For routine pediatric abdomen PCCT, start with Quantum mode at 70 kV and low-keV VMI rather than defaulting to QuantumPlus. Based on their phantom work, the authors favor the lowest tube voltage as the ALARA-oriented starting point, and for their standard protocol they selected 70 kV Quantum mode with VMI 53 keV. Optimize…
-
(2026) Hegmanns | PCCT vs EID-CT in Obesity
In 51 patients with obesity (BMI ≥30 kg/m²), intraindividual comparisons of thoracic and abdominal CT demonstrated superior image quality and lower radiation exposure with PCCT compared with conventional EID-CT. PCCT significantly reduced radiation dose: CTDIvol 11.99 vs 17.05 mGy (−30%) and DLP 526.25 vs 709.32 mGy·cm…
-
(2022) Lennartz | Virtual unenhanced images show limitations for adrenal adenoma characterization
PCCT-derived virtual unenhanced (VUE) images showed no significant difference in mean attenuation compared with true unenhanced (TUE) images for adrenal adenomas (9.2 ± 8.6 HU vs 7.3 ± 8.4 HU; P = .07). Despite similar mean values, individual VUE measurements showed both over- and underestimation, resulting in 10 of 36…
-
(2025) Zhang | UHR PCCT versus 40-keV VMI for abdominal artery visualization
This prospective study compared 0.2-mm ultra-high-resolution (UHR) PCCT with 1-mm standard-resolution 40-keV VMI for abdominal CTA in 94 participants, equally divided between full-dose and low-dose protocols. Standard-resolution 40-keV VMI provided significantly higher SNR and CNR than all UHR reconstructions. UHR…
-
(2024) Wang | Low-dose ultra-high-resolution PCCT for visceral artery CTA
Prospective study testing whether low-dose (LD) ultra-high-resolution (UHR) photon-counting CT can preserve diagnostic image quality for visceral artery CTA. The study included 57 patients in the full-dose group and 57 in the low-dose group. Each acquisition was reconstructed two ways: 0.2 mm / Bv48 / QIR 4 and 1 mm / Bv40…
-
(2023) Schade | Pseudoenhancement of renal cysts on PCCT
In 86 patients with 160 non-enhancing renal cysts, the study assessed how PCCT virtual monoenergetic image (VMI) energy level affects apparent renal cyst enhancement and therefore lesion classification. Low-keV VMI substantially increased apparent cyst attenuation: mean attenuation rose from 4 ± 11 HU at 190 keV to 36 ± 17…
-
(2026) Huang | UHR PCCT with low-keV improves preoperative pancreatic cancer assessment
Combining UHR PCCT with low-keV virtual monoenergetic imaging (VMI) improved visualization of pancreatic ductal adenocarcinoma (PDAC), peripancreatic vessels, and fine pancreatic anatomy; the study identified 0.4-mm slices at 45 keV as the best overall reconstruction strategy for preoperative assessment. 40-keV VMI…
-
(2022) Mergen | Virtual noncontrast abdominal imaging with PCCT
Virtual noncontrast (VNC) images reconstructed from contrast-enhanced PCCT provided accurate abdominal attenuation values compared with true noncontrast (TNC) imaging, supporting the potential to omit a separate TNC acquisition. Across all measurements, VNC attenuation errors were <5 HU in 76% and <10 HU in 95%; there was…
-
(2024) Khanungwanitkul | PCCT reveals pancreatic duct connection of a side-branch IPMN
This case report has been published in Radiology. Access publication Figure 1. Axial photon-counting CT images at (A) 0.4-mm, (C) 2-mm, and (D) 4-mm section thickness demonstrate an incidental hypoattenuating pancreatic lesion (arrowhead). The lesion communicates with the main pancreatic duct via a connection (arrow)…
-
(2024) Huflage | Submillisievert PCCT for urinary calculi detection
Retrospective single-center study of 507 patients undergoing unenhanced abdominal CT for suspected urinary calculi, comparing PCCT (n = 229) with optimized EID-CT (n = 278). PCCT achieved a 44% lower effective radiation dose than EID-CT: 0.79 mSv vs 1.39 mSv, while maintaining excellent diagnostic confidence. Despite the…
-
(2024) Rau | PCCT virtual non-contrast imaging accurately characterizes uncomplicated renal cysts
PCCT-derived virtual non-contrast (VNC) images provided renal cyst attenuation measurements comparable to true non-contrast imaging, with no significant difference across true non-contrast, arterial VNC, and portal-venous VNC (p = 0.301). In contrast, dual-energy EID-CT VNC showed a significant difference from true…
-
(2024) Lin | PCCT enables protocol-independent liver fat quantification
PCCT-derived fat fraction (CTFF) showed excellent agreement with MRI proton density fat fraction (PDFF), supporting quantitative liver fat assessment using standardized PCCT attenuation values. A single conversion formula—MRI PDFF (%) = −0.58 × CT (HU) + 43.1—remained accurate across 90, 120, and 140 kVp and both standard-…
-
(2023) Graafen | Soft Reconstruction Kernels Improve HCC Imaging on a PCCT
In a phantom and clinical study involving 24 patients with viable hepatocellular carcinoma (HCC), Graafen et al. evaluated reconstruction kernels for triple-phase photon-counting detector CT (PCD-CT), using 50-keV virtual monoenergetic images. Image quality was assessed quantitatively through contrast-to-noise ratio (CNR)…
-
(2024) Dillinger | Low-keV VMI from portal venous PCCT enhances abdominal arterial contrast
This retrospective study of 40 patients evaluated whether virtual monoenergetic images (VMI) reconstructed from portal venous PCCT could provide sufficient arterial vessel contrast and potentially serve as an alternative when a dedicated arterial phase is unavailable or suboptimal. 40-keV VMI produced the highest arterial…
-
(2022) Decker | PCCT enables anemia detection from routine contrast-enhanced CT
This proof-of-concept study demonstrates that spectral PCCT data from routine contrast-enhanced oncologic staging examinations can be used to detect and quantify anemia by measuring blood-pool attenuation on virtual non-contrast (VNC) images. Mean blood-pool VNC attenuation showed a strong positive correlation with…
-
(2024) Dane | Pancreatic cyst prevalence and detection with PCCT compared with EID-CT
In a retrospective single-center study of 2,494 patients, photon-counting CT (PCCT) demonstrated a significantly higher pancreatic cyst detection rate than conventional energy-integrating detector CT (EIDCT): 4.9% (49/1,009) vs 3.0% (44/1,485), p = 0.017. The largest difference occurred with CT angiography, where cysts…
-
(2024) Bette | PCCT virtual non-contrast imaging differentiates adrenal adenomas from metastases
PCCT virtual non-contrast (VNC) reconstructions differentiated adrenal adenomas from metastases on portal-venous contrast-enhanced imaging, with adenomas showing significantly lower attenuation on both conventional VNC (VNCConv) and calcium-preserving PureCalcium VNC (VNCPC) (p = 0.007). VNC nevertheless systematically…
-
(2022) Bette | Low-keV PCCT improves conspicuity of hypovascularized liver metastases
PCCT virtual monoenergetic imaging (VMI) improved visualization of hypovascularized liver metastases, with the greatest lesion conspicuity and CNR achieved at the low end of the keV spectrum. At 40 keV, tumor-to-liver ratio (TLR) reached 0.37, while CNR peaked at 6.88. Compared with conventional EID-CT, PCCT showed lower…
-
(2025) Becker | Spectral PCCT differentiates hyperdense vascular from non-vascular renal lesions
This pilot study demonstrates that spectral information from photon-counting detector CT (PCD-CT) can help differentiate hyperdense renal cysts from vascular renal lesions when conventional portal-venous images are indeterminate. In 26 patients with 42 hyperdense renal lesions, attenuation on standard 70-keV VMI…
Anstehende Veranstaltungen
-
NAEOTOM Connect at RSNA 2026
Let's Connect at RSNA 2026! Let's make RSNA 2026 an opportunity to learn, connect, and inspire one another! Whether you're attending to discover new clinical insights, share your experiences with photon-counting CT, or reconnect with colleagues, we look forward to bringing the NAEOTOM Connect community together. Join the…
-
Masterclass photon-counting CT course
This class is taught by Siemens Healthineers experts developing and integrating the technology since more than a decade. Starting off with the theory, you will deep dive into the complete imaging chain and complement your learning with practical experiences on clinical focus areas and best-practices. Customer quotes: "Very…
-
Hands-on PCCT in Rotterdam at Erasmus MC
Learn from one of our NAEOTOM Alpha ambassadors, Erasmus MC, how to leverage the full power of photon-counting CT in practice. From live patient scanning to protocols and post-processing - this class will make you ready for the transition or support in your daily routine, or get an advanced user of NAEOTOM Alpha. "When…
-
Spectral imaging with PCCT course (Date tbd)
The date hasn't been fixed, yet. Tag yourself as interested to receive updates. "We came expecting a spectral image reading class, but what we experienced was much more: a great opportunity to connect, exchange, and be part of the photon-counting CT community." "The combination of expert lectures, practical cases and open…