Cardiac PCCT integrated directly into the acute stroke CT workflow achieved 96% diagnostic accuracy, 90% sensitivity, 97% specificity, and a particularly strong 99% negative predictive value for left atrial appendage (LAA) thrombus compared with TEE.
The cardiac acquisition was added immediately after stroke CTA with only an ~1.5-s scan duration and an 8-s delay after CTA, without measurable delay to acute revascularization treatment. Prospective ECG triggering was targeted at 35% of the R–R interval, although successful gating was achieved in only about 60% of patients.
Diagnostic performance was numerically better with ECG-gating (AUC 0.99) than without gating (AUC 0.89), supporting ECG synchronization as an important technical optimization. The study nevertheless positions cardiac PCD-CT primarily as a strong rule-out tool for LAA thrombus, complementing rather than replacing echocardiography.