VNC-based coronary calcium scoring on PCCT showed strong correlation with TNC scoring, but systematically underestimated Agatston scores by about 11% overall.
Agreement in CAC severity category was 80%, but around 11% of patients were misclassified, which could potentially affect treatment decisions.
Accuracy was best in patients with BMI ≤28 kg/m² and heart rate ≤69 bpm, while underestimation increased in obese and tachycardic patients.
Access publication
Fig. 1. Demonstration of coronary artery calcium scoring (CACS) for two examples. Images show maximum intensity projections of axial slices for true non-contrast (TNC) and virtual non-contrast (VNC) reconstructions. Voxels with CT-values exceeding 130 HU are considered to represent calcifications (marked purple) and can be allocated semi-manually to single coronary arteries: green = left main artery (LM), yellow = left anterior descending artery (LAD), blue = circumflex artery (CX), red = right coronary artery (RCA).
Source: Becker et al., International Journal of Cardiovascular Imaging, 2025, CC BY 4.0
Protocol
Detail | Value |
|---|
Scanner | NAEOTOM Alpha |
Scan area | Heart |
Scan protocol | True non-contrast scan + coronary CTA |
Scan mode | TNC high-pitch spiral | CTA low-pitch ECG-triggered spiral |
Acquisition mode | Quantum Plus |
Tube voltage [kV] | 120 |
Effective mAs [mAs] | n.a. |
IQ level | 19 TNC | 50 CTA |
Dose modulation | n.a. |
CTDIvol [mGy] | 1.4 (unenhanced) | 28.3 (CTA) |
DLP [mGy*cm] | 27.9 | 437.5 |
Rotation time [s] | n.a. |
Pitch | High-pitch TNC | low-pitch CTA |
Matrix | 512 |
Slice collimation [mm] | 144 × 0.4 |
Slice width [mm] | 3.0 |
Reconstruction increment [mm] | 1.5 |
Reconstruction kernel | Qr36 |
keV level [keV] | 70 |
Iterative reconstruction | QIR 3 |
Reconstruction type | TNC from unenhanced scan; calcium-preserving VNC from CTA |
Energy thresholds [keV] | 20, 35, 65, 70 |
Contrast agent | Ultravist, iopromide 300 mgI/mL |
Contrast protocol | Triphasic bolus tracking protocol |
Contrast volume | 60 mL contrast + 30 mL contrast/30 mL saline mixture + 20 mL saline |
Flow rate | 5 mL/s in all phases |
Patient preparation | No beta-blockers administered |
Scoring threshold | 130 HU |
Software for scoring | Syngo.via VB60A |
Becker J, Decker JA, Bette S, et al. Impact of BMI, heart rhythm, and heart rate on photon-counting detector virtual coronary calcium scoring. Int J Cardiovasc Imaging. 2025;41:1331–1339.