For routine pediatric abdomen PCCT, start with Quantum mode at 70 kV and low-keV VMI rather than defaulting to QuantumPlus. Based on their phantom work, the authors favor the lowest tube voltage as the ALARA-oriented starting point, and for their standard protocol they selected 70 kV Quantum mode with VMI 53 keV.
Optimize image quality first through reconstruction settings before increasing acquisition dose. The authors recommend first adjusting QIR strength, kernel choice, and windowing if noise is too high, and only then considering higher IQ level, DAC changes, or QuantumPlus mode.
Use high-pitch FLASH selectively, not routinely. The authors found FLASH shortened scan time but slightly increased noise and dose, so they recommend it mainly for children who cannot hold their breath long enough for a standard scan.