The study evaluates a low-dose high-resolution PCCT lung protocol in a clinical cohort, comparing it directly to standard energy-integrating CT under routine conditions.
Image quality was significantly improved with PCCT, particularly when using quantum iterative reconstruction, which reduced noise and enhanced sharpness of pulmonary structures.
At the same time, the protocol achieved a meaningful reduction in radiation dose, demonstrating that optimized low-dose PCCT can be implemented in clinical practice without compromising diagnostic performance.
Access publication
Figure 1. Images of the whole lung (a–c) and enlarged image sections (d–f) of the same patient: (a,d) were created with the EID-CT scanner; (b,e) were created with a PCCT scanner without an iterative reconstruction algorithm (QIR−); and (c,f) are PCCT images reconstructed with an iterative reconstruction algorithm (QIR+).
Source: Woeltjen et al., Diagnostics, 2022, CC BY 4.0
Protocol
Scanner | NAEOTOM Alpha |
|---|
Scan area | Thorax |
Scan mode | Quantum Plus |
Scan direction | n.a. |
Tube voltage [kV] | 120 |
Effective mAs [mAs] | n.a. |
IQ level | 70 |
Dose modulation | CARE Dose4D |
CTDIvol [mGy] | 2.5 ± 1.1 |
DLP [mGy*cm] | 85.9 ± 36.1 |
Rotation time [s] | 0.25 |
Pitch | 1.2 |
Matrix | n.a. |
Slice collimation [mm] | 144 × 0.40 |
Slice width [mm] | 1.0 |
Reconstruction increment [mm] | n.a. |
Reconstruction kernel | QIR 2 |
Woeltjen et al. Low-Dose High-Resolution Photon-Counting CT of the Lung: A Clinical Study. Diagnostics. 2022;12(6):1441.