In 71 post-TAVI PCCT examinations from 63 patients, time-resolved cardiac PCCT enabled calculation of prosthetic-valve hemodynamics from left-ventricular volumetry and direct valve-orifice planimetry. CT-derived mean and maximum transvalvular gradients correlated strongly with echocardiography (ρ = 0.85 for both), while corrected valve area showed a correlation of ρ = 0.73.
The most clinically useful parameter was the mean transvalvular gradient (Pmean). Using echocardiographic Pmean ≥20 mmHg as the reference for potential valve dysfunction, PCCT achieved AUC 0.94, sensitivity 88.9%, specificity 88.6%, and overall accuracy 88.7%. The independently derived optimal PCCT cutoff was almost identical at 20.4 mmHg.
PCCT-derived Pmean showed much better agreement than Pmax: mean Pmean bias was only 1.6 mmHg, whereas Pmax was systematically higher and showed wider limits of agreement. This suggests that a post-TAVI PCCT already obtained for structural assessment could potentially provide opportunistic functional valve assessment, although the retrospective single-center workflow still requires external validation.