Ultra-high-resolution (UHR) PCCT provided better subjective image quality than standard-resolution (SR) PCCT for femoral stent CTA, while stent-lumen visibility itself remained comparable between modes.
The study used a perfused human cadaveric upper-leg model with 7 superficial femoral artery stents. SR and UHR acquisitions were dose-matched at 120 kVp and CTDIvol 3, 5, and 10 mGy, with four vascular reconstruction kernels evaluated.
The advantage of UHR became particularly apparent with sharp/ultra-sharp reconstruction: CNR was significantly higher with UHR for BV60 and BV76, and readers preferred UHR image quality across all evaluated kernels, with BV76 ranked highest.
Importantly, the UHR benefit was largest at low radiation dose. With BV76, the relative CNR improvement was 48% at low dose vs 36% at high dose. A 3-mGy UHR acquisition achieved CNR comparable to a 10-mGy SR acquisition (8.0 ± 0.6 vs 9.1 ± 1.1; p > .760).