This phantom study compared photon-counting CT (PCCT), dual-energy CT (DECT), and single-energy CT (SECT) for proton stopping-power ratio (SPR) estimation, specifically testing how phantom size and insert position affect accuracy through beam hardening. Two SPR methods were evaluated: a DECT-based Saito method and a conventional Hounsfield look-up table (HLUT) method.
The principal finding was that PCCT combined with VMI-based HLUT estimation was the most stable approach, particularly in larger phantoms. For PCCT, non-bone SPR deviations with HLUT were essentially within ±1%, while peripheral errors became substantially larger for DECT and SECT.
In the most challenging 40-cm phantom, PCCT + HLUT limited peripheral deviation to +2.1% for cortical bone, whereas deviations reached −5.5% for DECT/SECT configurations. Overall, the authors conclude that PCCT-VMI + HLUT is less susceptible to size- and position-dependent beam hardening than DECT or SECT.