<?xml version="1.0" encoding="utf-8"?>
<rss version="2.0"
    xmlns:content="http://purl.org/rss/1.0/modules/content/"
    xmlns:dc="http://purl.org/dc/elements/1.1/"
    xmlns:atom="http://www.w3.org/2005/Atom">
    <channel>
        <title>QuantumPlus — NAEOTOM Connect</title>
        <link>https://connect.siemens-healthineers.com/de/</link>
        <pubDate>Mon, 23 Nov 2026 04:15:40 +0000</pubDate>
        <language>de</language>
            <description>QuantumPlus — NAEOTOM Connect</description>
    <atom:link href="https://connect.siemens-healthineers.com/de/discussions/tagged/quantumplus/feed.rss" rel="self" type="application/rss+xml"/>
    <item>
        <title>(2026) Darvizeh | PCCT in oral and maxillofacial imaging</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/640/2026-darvizeh-pcct-in-oral-and-maxillofacial-imaging</link>
        <pubDate>Sat, 21 Nov 2026 07:13:03 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">640@/de/discussions</guid>
        <description><![CDATA[<div><div><p>This narrative review identifies <strong>high spatial resolution, intrinsic spectral imaging, metal-artifact reduction, and potential dose efficiency</strong> as the main advantages of PCCT for oral and maxillofacial imaging. The strongest technical evidence concerns visualization of <strong>fine endodontic anatomy, peri-implant bone, maxillofacial fractures, and sinonasal structures</strong>, but much of the literature remains phantom, ex vivo, or early clinical evidence.</p><p> </p><p>PCCT appears particularly promising when <strong>metal artifacts or very small osseous structures</strong> limit conventional imaging. Artificial accessory canals down to <strong>200 µm</strong> were reliably depicted in one PCCT phantom study, and high-keV VMI around <strong>120–130 keV</strong> substantially reduced titanium implant artifacts while preserving peri-implant detail. However, optimized high-resolution CBCT can equal or outperform PCCT for some tasks, including root-crack visualization. </p><p></p><p><strong>CBCT remains the preferred modality for most routine dental indications.</strong> PCCT dose performance is highly protocol- and task-dependent: some studies demonstrated approximately dose-matched or lower-dose PCCT, whereas another endodontic study reported <strong>0.85 mSv with PCCT versus 0.01–0.13 mSv with CBCT</strong>. The authors therefore position PCCT as a complementary technique for complex cases rather than a routine replacement for CBCT. </p></div></div><p><strong></strong><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fpmc.ncbi.nlm.nih.gov%2Farticles%2FPMC13761622%2F" rel="nofollow noopener ugc" role="button">Access publication</a></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/78UD3PFSMOG9\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:1207975,&quot;width&quot;:1423,&quot;height&quot;:936,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F78UD3PFSMOG9%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:2115,&quot;dateInserted&quot;:&quot;2026-11-19T07:10:26+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F78UD3PFSMOG9%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/78UD3PFSMOG9/image.png" alt="image.png" height="936" width="1423" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/78UD3PFSMOG9/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/78UD3PFSMOG9/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/78UD3PFSMOG9/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/78UD3PFSMOG9/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/78UD3PFSMOG9/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/78UD3PFSMOG9/image.png 2000w, https://us.v-cdn.net/6038601/uploads/78UD3PFSMOG9/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 2. </strong>High-resolution dental imaging with photon-counting CT. (A) Three-dimensional cinematic rendering of the dentition; (B) longitudinal multiplanar reconstruction of the upper and lower jaws; and (C) detailed cross-sectional visualization of a single tooth, its root, and the surrounding alveolar bone. The images demonstrate high-resolution dental imaging achievable with clinical photon-counting CT, with a displayed in-plane resolution of approximately 100 μm. Reproduced from Ref. [1].</p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner used in representative studies</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Primary applications</strong></p></td><td><p>Dental/endodontic imaging, peri-implant assessment, maxillofacial trauma, sinonasal imaging</p></td></tr><tr><td><p><strong>Acquisition mode</strong></p></td><td><p>Task-dependent; UHR frequently used for fine osseous and dental structures</p></td></tr><tr><td><p><strong>Spatial resolution</strong></p></td><td><p>Voxel sizes approaching ~100 µm reported in selected UHR applications</p></td></tr><tr><td><p><strong>Spectral reconstruction</strong></p></td><td><p>VMI and material decomposition</p></td></tr><tr><td><p><strong>Metal artifact reduction</strong></p></td><td><p>High-keV VMI; iMAR evaluated in selected protocols</p></td></tr><tr><td><p><strong>Implant VMI range</strong></p></td><td><p>70–190 keV evaluated</p></td></tr><tr><td><p><strong>Preferred implant VMI</strong></p></td><td><p>Approximately 120–130 keV</p></td></tr><tr><td><p><strong>Endodontic PCCT CTDI [mGy]</strong></p></td><td><p>6.01 in Rau et al.</p></td></tr><tr><td><p><strong>Endodontic PCCT DLP [mGy·cm]</strong></p></td><td><p>52.43 in Rau et al.</p></td></tr><tr><td><p><strong>Peri-implant PCCT CTDI [mGy]</strong></p></td><td><p>4 in Ruetters et al.; 1 in Sawall et al.</p></td></tr><tr><td><p><strong>Implant-artifact PCCT CTDIvol [mGy]</strong></p></td><td><p>3.28 in Al-Haj Husain et al.</p></td></tr><tr><td><p><strong>Implant-artifact DLP [mGy·cm]</strong></p></td><td><p>45.9</p></td></tr><tr><td><p><strong>Estimated effective dose</strong></p></td><td><p>92 µSv in implant-artifact study</p></td></tr><tr><td><p><strong>Matched-dose endodontic PCCT</strong></p></td><td><p>13.1 / 75 / 102 µSv for low / standard / high dose</p></td></tr><tr><td><p><strong>Sinonasal ultra-low-dose imaging</strong></p></td><td><p>Diagnostic imaging reported at exposures as low as 0.08 mGy</p></td></tr><tr><td><p><strong>Protocol principle</strong></p></td><td><p>Task-specific optimization of FOV, spatial resolution, reconstruction, and dose; no standardized indication-specific PCCT protocol currently exists</p></td></tr></table></div><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Application</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Accessory root canals</strong></p></td><td><p>PCCT reliably depicted artificial canals down to <strong>200 µm</strong></p></td></tr><tr><td><p><strong>Endodontic diagnostic accuracy</strong></p></td><td><p><strong>80–88% PCCT</strong> vs <strong>75–88% CBCT</strong> in approximately dose-matched ex vivo testing</p></td></tr><tr><td><p><strong>Peri-implant bone measurement</strong></p></td><td><p>Mean deviation <strong>0.06 ± 0.08 mm PCCT</strong> vs <strong>0.39 ± 0.34 mm CBCT</strong></p></td></tr><tr><td><p><strong>Implant artifacts</strong></p></td><td><p><strong>120–130 keV VMI</strong> provided best artifact reduction/image-quality balance</p></td></tr><tr><td><p><strong>Dental dose efficiency</strong></p></td><td><p>One study: PCCT <strong>1 mGy CTDI</strong> vs DVT <strong>4 mGy</strong> with better dose-normalized image quality</p></td></tr><tr><td><p><strong>Pediatric temporal bone</strong></p></td><td><p>~<strong>43% overall dose reduction</strong> vs EID-CT; &gt;70% in children &lt;6 years</p></td></tr><tr><td><p><strong>Sinus imaging</strong></p></td><td><p>Ultra-low-dose PCCT reported superior image quality with mean effective dose about <strong>one-quarter of CBCT</strong> in one study</p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Darvizeh A, Darvizeh F, Calegari L, Perera Molligoda Arachchige AS, Maspero C. <em>Photon-Counting CT in Oral and Maxillofacial Imaging: Clinical Applications, Dose Efficiency, and Future Perspectives.</em> <strong>Diagnostics.</strong> 2026;16(21):3617. doi:10.3390/diagnostics16213617</p>]]>
        </description>
    </item>
    <item>
        <title>(2026) Chen | Phase-specific reconstruction optimizes low-dose liver VPCT on PCCT</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/642/2026-chen-phase-specific-reconstruction-optimizes-low-dose-liver-vpct-on-pcct</link>
        <pubDate>Fri, 20 Nov 2026 07:03:19 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">642@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In <strong>62 patients with liver disease</strong>, PCCT-based liver volume perfusion CT (VPCT) enabled a one-stop workflow combining perfusion with morphologic assessment. Compared with a simulated conventional four-phase CT plus perfusion workflow, effective dose decreased from <strong>31.9 ± 8.5 to 25.9 ± 6.7 mSv</strong>, an approximately <strong>18.8% reduction</strong>, while perfusion-derived morphologic images generally remained diagnostically acceptable. </p><p></p><p>The optimal reconstruction depended strongly on imaging phase. <strong>Fused noncontrast images</strong> provided the best objective noncontrast image quality, whereas <strong>40-keV VMI</strong> maximized arterial vascular enhancement and CNR; for routine arterial assessment, however, <strong>60-keV VMI and T3D</strong> achieved the highest overall subjective quality because 40-keV images were noisier. </p><p></p><p>In the portal venous phase, <strong>60-keV VMI and T3D</strong> provided the best overall subjective quality, while fused images produced the lowest noise. <strong>40-keV VMI was subdiagnostic for general portal venous morphologic assessment</strong>, illustrating why a single reconstruction cannot be applied optimally across all liver phases. </p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fwww.academicradiology.org%2Farticle%2FS1076-6332%2826%2900859-7%2Ffulltext" rel="nofollow noopener ugc" role="button">Access publication</a></p><p><strong></strong></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/819KZO7NVV37\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:3284824,&quot;width&quot;:2994,&quot;height&quot;:1529,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F819KZO7NVV37%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:2117,&quot;dateInserted&quot;:&quot;2026-11-19T07:19:24+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F819KZO7NVV37%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/819KZO7NVV37/image.png" alt="image.png" height="1529" width="2994" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/819KZO7NVV37/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/819KZO7NVV37/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/819KZO7NVV37/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/819KZO7NVV37/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/819KZO7NVV37/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/819KZO7NVV37/image.png 2000w, https://us.v-cdn.net/6038601/uploads/819KZO7NVV37/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 7 </strong>A female patient with hepatic hemangioma, 61 years old. 3a-3d: single-phase arterial phase images of perfusion derived from VMI reconstructions at 40 keV, 50 keV, 60 keV, and 70 keV; 3e is the fused arterial phase image; 3f is the arterial phase image of T3D (red arrowhead). </p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Liver, diaphragmatic dome to inferior hepatic margin</p></td></tr><tr><td><p><strong>Patient position</strong></p></td><td><p>Supine, both arms raised</p></td></tr><tr><td><p><strong>Acquisition protocol</strong></p></td><td><p>True noncontrast + dynamic VPCT + low-dose delayed phase</p></td></tr><tr><td><p><strong>TNC scan mode</strong></p></td><td><p>QuantumPlus</p></td></tr><tr><td><p><strong>TNC tube voltage</strong></p></td><td><p>140 kV</p></td></tr><tr><td><p><strong>TNC IQ level</strong></p></td><td><p>130</p></td></tr><tr><td><p><strong>VPCT scan mode</strong></p></td><td><p>Quantum</p></td></tr><tr><td><p><strong>VPCT tube voltage</strong></p></td><td><p>90 kV</p></td></tr><tr><td><p><strong>VPCT IQ level</strong></p></td><td><p>70</p></td></tr><tr><td><p><strong>Delayed scan mode</strong></p></td><td><p>QuantumPlus</p></td></tr><tr><td><p><strong>Delayed tube voltage</strong></p></td><td><p>120 kV</p></td></tr><tr><td><p><strong>Delayed IQ level</strong></p></td><td><p>90</p></td></tr><tr><td><p><strong>Dose modulation</strong></p></td><td><p>CARE Dose4D</p></td></tr><tr><td><p><strong>VPCT CTDIvol [mGy]</strong></p></td><td><p>59.7 ± 13.3</p></td></tr><tr><td><p><strong>VPCT DLP [mGy·cm]</strong></p></td><td><p>1431.3 ± 372.1</p></td></tr><tr><td><p><strong>VPCT effective dose [mSv]</strong></p></td><td><p>21.8 ± 5.9</p></td></tr><tr><td><p><strong>VPCT temporal interval</strong></p></td><td><p>Approximately 1.8 s during continuous acquisition</p></td></tr><tr><td><p><strong>VPCT acquisition duration [s]</strong></p></td><td><p>49.3, beginning 7 s after contrast injection</p></td></tr><tr><td><p><strong>Number of VPCT cycles</strong></p></td><td><p>21–28</p></td></tr><tr><td><p><strong>Z-axis coverage [cm]</strong></p></td><td><p>16–26, adapted to patient size</p></td></tr><tr><td><p><strong>Rotation time [s]</strong></p></td><td><p>0.25</p></td></tr><tr><td><p><strong>Pitch</strong></p></td><td><p>1.0</p></td></tr><tr><td><p><strong>Detector slice collimation</strong></p></td><td><p>144 × 0.4 mm</p></td></tr><tr><td><p><strong>Reconstruction slice thickness [mm]</strong></p></td><td><p>1.0</p></td></tr><tr><td><p><strong>Reconstruction increment [mm]</strong></p></td><td><p>0.7</p></td></tr><tr><td><p><strong>Kernel</strong></p></td><td><p>Br40</p></td></tr><tr><td><p><strong>Iterative reconstruction</strong></p></td><td><p>QIR 4</p></td></tr><tr><td><p><strong>Spectral reconstruction</strong></p></td><td><p>T3D; VMI at 40, 50, 60, and 70 keV; fused images</p></td></tr><tr><td><p><strong>Contrast</strong></p></td><td><p>Iohexol 300 mg I/mL</p></td></tr><tr><td><p><strong>Contrast volume [mL]</strong></p></td><td><p>60</p></td></tr><tr><td><p><strong>Injection rate [mL/s]</strong></p></td><td><p>5</p></td></tr><tr><td><p><strong>Saline flush [mL]</strong></p></td><td><p>50</p></td></tr><tr><td><p><strong>Post-processing</strong></p></td><td><p>syngo.via VB60, CT Dynamic Angio</p></td></tr><tr><td><p><strong>Motion/noise processing</strong></p></td><td><p>Automatic motion correction and 4D noise reduction</p></td></tr><tr><td><p><strong>Perfusion phase selection</strong></p></td><td><p>First acquisition = noncontrast; aortic peak = arterial; portal venous peak = portal venous</p></td></tr><tr><td><p><strong>Fused noncontrast</strong></p></td><td><p>Average of first 3 consecutive VPCT time points</p></td></tr><tr><td><p><strong>Fused arterial phase</strong></p></td><td><p>Aortic peak averaged with immediately preceding and following phases</p></td></tr><tr><td><p><strong>Fused portal venous phase</strong></p></td><td><p>Portal venous peak averaged with immediately preceding and following phases</p></td></tr><tr><td><p><strong>Fusion algorithm</strong></p></td><td><p>Temporal Average (tAVG)</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p><strong>62 patients</strong></p></td></tr><tr><td><p><strong>Effective dose</strong></p></td><td><p>One-stop <strong>25.9 ± 6.7 mSv</strong> vs simulated conventional + VPCT <strong>31.9 ± 8.5 mSv</strong></p></td></tr><tr><td><p><strong>Fused noncontrast</strong></p></td><td><p>Noise <strong>7.8 ± 2.9 HU</strong>, SNR <strong>8.6 ± 3.0</strong>, CNR <strong>1.5 ± 1.2</strong></p></td></tr><tr><td><p><strong>Arterial 40-keV VMI</strong></p></td><td><p>Highest vessel SNR <strong>107.1 ± 49.0</strong> and vessel CNR <strong>101.8 ± 47.3</strong></p></td></tr><tr><td><p><strong>Fused arterial phase</strong></p></td><td><p>Lowest noise <strong>7.5 ± 2.7 HU</strong>; vessel CNR <strong>88.4 ± 42.6</strong></p></td></tr><tr><td><p><strong>Portal venous phase</strong></p></td><td><p>Fused images lowest noise <strong>8.6 ± 2.8 HU</strong>; 60-keV VMI/T3D highest subjective overall quality</p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Chen Y, Ma W, Ma B, Ma X, Chen L, Liu Y, Wang A, Chen P, Wang Z. <em>Low-Dose PCCT-Based Liver Volume Perfusion CT with Phase-Specific Reconstruction Optimization for Morphologic.</em> <strong>Academic Radiology.</strong> Published online November 10, 2026. doi:10.1016/j.acra.2026.10.024.</p>]]>
        </description>
    </item>
    <item>
        <title>(2026) Herold | UHR PCCT reconstruction settings for rectal cancer imaging</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/639/2026-herold-uhr-pcct-reconstruction-settings-for-rectal-cancer-imaging</link>
        <pubDate>Thu, 19 Nov 2026 07:08:19 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">639@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In this prospective pilot study of 19 patients, UHR PCCT image quality in rectal cancer was strongly influenced by both slice thickness and monoenergetic level. Thicker slices increased SNR and reduced noise, while lower keV increased tumor-to-muscle CNR at the expense of higher noise; CNR ranged from 4.85 ± 2.43 at 3 mm/45 keV to 0.86 ± 1.04 at 0.6 mm/70 keV. </p><p></p><p>Qualitatively, higher monoenergetic levels improved overall image quality and mesorectal fascia sharpness, whereas lower keV favored lesion conspicuity. The most frequently favored pooled reconstruction was 3 mm at 60 keV (68.0%), closely followed by 2 mm at 70 keV (66.2%); overlapping confidence intervals mean that neither setting can be considered definitively optimal. </p><p></p><p>The study was designed for image-quality optimization rather than diagnostic staging. The additional pelvic UHR acquisition roughly doubled CTDIvol compared with the standard portal venous phase and added about 5.1 mSv, so the authors emphasize that future studies must demonstrate a diagnostic benefit before this additional dose can be justified.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fpmc.ncbi.nlm.nih.gov%2Farticles%2FPMC13762976%2F" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/2JJ9AWTZOGNO\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:761237,&quot;width&quot;:1233,&quot;height&quot;:825,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F2JJ9AWTZOGNO%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:2114,&quot;dateInserted&quot;:&quot;2026-11-19T07:07:32+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F2JJ9AWTZOGNO%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/2JJ9AWTZOGNO/image.png" alt="image.png" height="825" width="1233" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/2JJ9AWTZOGNO/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/2JJ9AWTZOGNO/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/2JJ9AWTZOGNO/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/2JJ9AWTZOGNO/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/2JJ9AWTZOGNO/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/2JJ9AWTZOGNO/image.png 2000w, https://us.v-cdn.net/6038601/uploads/2JJ9AWTZOGNO/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 4. </strong>Representative ultra-high-resolution photon-counting CT images in a patient with adenocarcinoma of the lower rectal third (arrows). (A–D) Ultra-high-resolution reconstructions at 3 mm slice thickness and 45, 50, 60 and 70 keV. (E) Standard reference reconstruction (QuantumPlus, 3 mm, 60 keV). Tumor-to-muscle contrast is highest in (A,B) and decreases toward (D), while perceived noise decreases in the same direction; in this example, the mesorectal fascia appears more sharply delineated in (C,D) than in the reference image (E), which was acquired in a different contrast phase. keV, kiloelectron volt. </p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Abdomen/pelvis staging CT plus dedicated pelvic UHR acquisition</p></td></tr><tr><td><p><strong>Standard scan mode</strong></p></td><td><p>QuantumPlus</p></td></tr><tr><td><p><strong>UHR scan mode</strong></p></td><td><p>HighResUltra</p></td></tr><tr><td><p><strong>Scan direction</strong></p></td><td><p>Cranio-caudal</p></td></tr><tr><td><p><strong>Tube voltage [kVp]</strong></p></td><td><p>120</p></td></tr><tr><td><p><strong>Quality reference [mAs]</strong></p></td><td><p>Portal venous: 121; UHR pelvis: 225</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>Portal venous: 170; UHR pelvis: 250</p></td></tr><tr><td><p><strong>CTDIvol [mGy]</strong></p></td><td><p>Portal venous: 8.4 ± 2.5; UHR pelvis: 18.3 ± 2.9</p></td></tr><tr><td><p><strong>DLP [mGy·cm]</strong></p></td><td><p>Portal venous: 556.4 ± 167.9; UHR pelvis: 343.0 ± 144.9</p></td></tr><tr><td><p><strong>Rotation time [ms]</strong></p></td><td><p>500</p></td></tr><tr><td><p><strong>Pitch</strong></p></td><td><p>0.80</p></td></tr><tr><td><p><strong>Detector slice collimation</strong></p></td><td><p>Portal venous: 144 × 0.4 mm; UHR pelvis: 120 × 0.2 mm</p></td></tr><tr><td><p><strong>Matrix</strong></p></td><td><p>Portal venous: 512; UHR pelvis: 768</p></td></tr><tr><td><p><strong>FOV</strong></p></td><td><p>Individual</p></td></tr><tr><td><p><strong>Reconstruction slice thickness [mm]</strong></p></td><td><p>Reference: 3; UHR: 0.6, 1, 2, 3</p></td></tr><tr><td><p><strong>Reconstruction increment [mm]</strong></p></td><td><p>Portal venous: 0.8; UHR: 0.4/variable</p></td></tr><tr><td><p><strong>Kernel</strong></p></td><td><p>Portal venous: Br40; UHR: Br48</p></td></tr><tr><td><p><strong>Iterative reconstruction</strong></p></td><td><p>QIR 3</p></td></tr><tr><td><p><strong>Monoenergetic reconstruction</strong></p></td><td><p>UHR: 45, 50, 60, 70 keV</p></td></tr><tr><td><p><strong>Reference reconstruction</strong></p></td><td><p>QuantumPlus, 3 mm, 60 keV</p></td></tr><tr><td><p><strong>Contrast</strong></p></td><td><p>Iomeron 400</p></td></tr><tr><td><p><strong>Contrast volume</strong></p></td><td><p>1 mL/kg body weight</p></td></tr><tr><td><p><strong>Portal venous phase</strong></p></td><td><p>Absolute delay 77 s</p></td></tr><tr><td><p><strong>Dedicated UHR pelvic phase</strong></p></td><td><p>Absolute delay 40 s</p></td></tr><tr><td><p><strong>Breathing instruction</strong></p></td><td><p>Inspiration</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p><strong>19 patients</strong></p></td></tr><tr><td><p><strong>Highest CNR</strong></p></td><td><p><strong>3 mm / 45 keV: 4.85 ± 2.43</strong></p></td></tr><tr><td><p><strong>Lowest CNR</strong></p></td><td><p><strong>0.6 mm / 70 keV: 0.86 ± 1.04</strong></p></td></tr><tr><td><p><strong>Best pooled qualitative rating</strong></p></td><td><p><strong>3 mm / 60 keV: 68.0% favorable</strong></p></td></tr><tr><td><p><strong>Alternative leading setting</strong></p></td><td><p><strong>2 mm / 70 keV: 66.2% favorable</strong></p></td></tr><tr><td><p><strong>MRF sharpness</strong></p></td><td><p>Favorable ratings <strong>79.8% at 60 keV</strong> and <strong>86.0% at 70 keV</strong></p></td></tr><tr><td><p><strong>UHR dose penalty</strong></p></td><td><p>Additional UHR acquisition: <strong>CTDIvol 18.3 ± 2.9 mGy</strong>, about <strong>5.1 mSv</strong> extra</p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Herold A, Strassl A, Beitzke D, Stelzer P, Wassipaul C, Dachs T, Nemeth B, Lindenlaub F, Ba-Ssalamah A, Laengle J, Bergmann M, Weber M, Tamandl D, Arnoldner MA. <em>Ultra-High-Resolution Imaging of Rectal Cancer with Photon-Counting Detector Computed Tomography: An Exploratory Pilot Study on the Impact of Slice Thickness and Monoenergetic Reconstruction Level on Image Quality.</em> <strong>Diagnostics.</strong> 2026;16(21):3697. doi:10.3390/diagnostics16213697.</p>]]>
        </description>
    </item>
    <item>
        <title>(2026) Liu | PCCT-derived ECV identifies cirrhosis and tracks functional severity</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/621/2026-liu-pcct-derived-ecv-identifies-cirrhosis-and-tracks-functional-severity</link>
        <pubDate>Thu, 12 Nov 2026 13:23:12 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">621@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In <strong>112 participants</strong>—69 patients with cirrhosis and 43 non-cirrhotic controls—PCCT-derived extracellular volume (ECV) was substantially higher in cirrhosis (<strong>29 ± 7% vs 20 ± 4%</strong>, <em>p</em> &lt; 0.001). Among multiple quantitative PCCT biomarkers, ECV was the <strong>only independent marker</strong> of cirrhosis in multivariable analysis (OR <strong>1.41</strong> per 1-percentage-point increase; 95% CI 1.16–1.72; <em>p</em> &lt; 0.001).</p><p></p><p>ECV also independently tracked <strong>functional severity</strong> across three established clinical systems: Child-Pugh (adjusted OR <strong>1.42</strong>), ALBI (adjusted OR <strong>1.33</strong>), and MELD-Na (adjusted OR <strong>1.38</strong>). This supports ECV as a structural imaging biomarker complementary to conventional liver-function scores rather than merely another perfusion parameter.</p><p></p><p>Diagnostic performance was good across tasks: AUC <strong>0.85</strong> for cirrhosis detection and <strong>0.85–0.88</strong> for severity stratification. The optimal ECV cutoff for cirrhosis was <strong>24.68%</strong> (71% sensitivity, 86% specificity), while higher thresholds were associated with more advanced functional impairment. </p></div></div><p><strong></strong><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fpmc.ncbi.nlm.nih.gov%2Farticles%2FPMC13743329%2F" rel="nofollow noopener ugc" role="button">Access publication</a></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/ZXUN54992DUY\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:898308,&quot;width&quot;:945,&quot;height&quot;:923,&quot;displaySize&quot;:&quot;medium&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FZXUN54992DUY%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:2062,&quot;dateInserted&quot;:&quot;2026-11-12T13:20:55+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FZXUN54992DUY%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/ZXUN54992DUY/image.png" alt="image.png" height="923" width="945" data-display-size="medium" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/ZXUN54992DUY/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/ZXUN54992DUY/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/ZXUN54992DUY/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/ZXUN54992DUY/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/ZXUN54992DUY/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/ZXUN54992DUY/image.png 2000w, https://us.v-cdn.net/6038601/uploads/ZXUN54992DUY/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 2. </strong>Representative example of ROI placement for ECV measurement on delayed-phase PCD-CT iodine maps. White ROIs were placed in the liver parenchyma (3 in the left lobe and 4 in the right lobe), and the blue ROI was placed in the abdominal aorta. ECV, extracellular volume; PCD-CT, photon counting detector computed tomography; ROI, region of interest. </p><p></p><h3 data-id="protocol">Protocol</h3><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p>Detail</p></th><th><p>Value</p></th></tr><tr><td><p>Scanner</p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p>Scan area</p></td><td><p>Abdomen / liver</p></td></tr><tr><td><p>Scan mode</p></td><td><p>QuantumPlus</p></td></tr><tr><td><p>Acquisition phases</p></td><td><p>Unenhanced, arterial, portal venous, delayed</p></td></tr><tr><td><p>Patient preparation</p></td><td><p>Fasting for 4–6 h; supine position with both arms raised</p></td></tr><tr><td><p>Tube voltage</p></td><td><p>Automated patient-specific kV selection with CARE keV</p></td></tr><tr><td><p>IQ level</p></td><td><p>145</p></td></tr><tr><td><p>Dose modulation</p></td><td><p>Automatic tube voltage and tube current modulation</p></td></tr><tr><td><p>CTDIvol [mGy]</p></td><td><p>Unenhanced 4.87 ± 0.76; arterial 4.87 ± 0.75; portal venous 4.87 ± 0.76; delayed 4.87 ± 0.76</p></td></tr><tr><td><p>DLP [mGy·cm]</p></td><td><p>Total four-phase examination: 646.8 ± 217.8</p></td></tr><tr><td><p>Rotation time [s]</p></td><td><p>0.5</p></td></tr><tr><td><p>Pitch</p></td><td><p>0.80</p></td></tr><tr><td><p>FOV [mm]</p></td><td><p>300–500, adjusted to body habitus</p></td></tr><tr><td><p>Detector slice collimation [mm]</p></td><td><p>Single collimation 0.40; total collimation 57.60</p></td></tr><tr><td><p>Matrix</p></td><td><p>512 × 512</p></td></tr><tr><td><p>Reconstruction slice thickness [mm]</p></td><td><p>1.0</p></td></tr><tr><td><p>Reconstruction increment [mm]</p></td><td><p>0.7</p></td></tr><tr><td><p>Kernel</p></td><td><p>Qr40</p></td></tr><tr><td><p>Iterative reconstruction</p></td><td><p>QIR 3</p></td></tr><tr><td><p>Spectral reconstruction</p></td><td><p>60-keV VMI; iodine water–iodine maps</p></td></tr><tr><td><p>Contrast</p></td><td><p>Ioversol 350 mg I/mL</p></td></tr><tr><td><p>Contrast volume</p></td><td><p>1.0–1.2 mL/kg</p></td></tr><tr><td><p>Injection rate [mL/s]</p></td><td><p>3.0–3.5</p></td></tr><tr><td><p>Saline flush [mL]</p></td><td><p>30, at same injection rate</p></td></tr><tr><td><p>Bolus trigger</p></td><td><p>Abdominal aorta at hepatic hilum; threshold 100 HU</p></td></tr><tr><td><p>Arterial phase</p></td><td><p>10 s after bolus trigger</p></td></tr><tr><td><p>Portal venous phase</p></td><td><p>25 s after completion of arterial phase</p></td></tr><tr><td><p>Delayed phase</p></td><td><p>3 min after start of contrast injection</p></td></tr><tr><td><p>Post-processing</p></td><td><p>syngo.via VB60B</p></td></tr><tr><td><p>ECV measurement levels</p></td><td><p>Hepatic hilum, 5 mm above, and 5 mm below</p></td></tr><tr><td><p>Liver ROIs</p></td><td><p>Seven 10-mm circular ROIs per level: 4 right lobe, 3 left lobe</p></td></tr><tr><td><p>Aortic ROI</p></td><td><p>One circular ROI per level in abdominal aorta</p></td></tr><tr><td><p>Total measurements</p></td><td><p>21 liver ROIs and 3 aortic ROIs per participant</p></td></tr><tr><td><p>ECV calculation</p></td><td><p>(1 − hematocrit) × (liver iodine concentration / aortic iodine concentration) × 100%</p></td></tr><tr><td><p>Hematocrit timing</p></td><td><p>Blood test within 1 week of CT</p></td></tr><tr><td><p>Additional quantitative outputs</p></td><td><p>Fat fraction, IC-PVP, IC-DP, nIC-PVP, nIC-DP, spectral slopes at 40–70 and 70–100 keV</p></td></tr></table></div><p></p><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p><strong>69 cirrhosis / 43 controls</strong></p></td></tr><tr><td><p><strong>ECV</strong></p></td><td><p><strong>29 ± 7% vs 20 ± 4%</strong>, <em>p</em> &lt; 0.001</p></td></tr><tr><td><p><strong>Independent marker of cirrhosis</strong></p></td><td><p>ECV OR <strong>1.41</strong> (95% CI 1.16–1.72), <em>p</em> &lt; 0.001</p></td></tr><tr><td><p><strong>Child-Pugh severity</strong></p></td><td><p>Adjusted OR <strong>1.42</strong> (95% CI 1.23–1.69)</p></td></tr><tr><td><p><strong>ALBI / MELD-Na severity</strong></p></td><td><p>Adjusted OR <strong>1.33</strong> / <strong>1.38</strong></p></td></tr><tr><td><p><strong>Cirrhosis detection</strong></p></td><td><p>AUC <strong>0.85</strong>; cutoff <strong>24.68%</strong>; sensitivity <strong>71%</strong>, specificity <strong>86%</strong></p></td></tr><tr><td><p><strong>Severity discrimination</strong></p></td><td><p>AUCs <strong>0.85–0.88</strong></p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Liu Y, Zhang Z, Wang T, Liu X, Li W, Liu S, Xu Z, Han B, Jiang H, Li J. <em>Photon-counting detector CT-derived extracellular volume fraction as a complementary biomarker for liver cirrhosis assessment: a retrospective diagnostic accuracy study.</em> <strong>Quantitative Imaging in Medicine and Surgery.</strong> 2026;16(11):856. doi:10.21037/qims-2026-1385. </p>]]>
        </description>
    </item>
    <item>
        <title>(2026) Kawasaki | 40-keV prone PCCT improves breast cancer conspicuity</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/556/2026-kawasaki-40-kev-prone-pcct-improves-breast-cancer-conspicuity</link>
        <pubDate>Tue, 27 Oct 2026 05:00:59 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">556@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In 22 women with surgically confirmed breast cancer, arterial-phase 40-keV VMI from prone PCCT achieved the highest lesion conspicuity, with median reader-averaged scores of 4.33 versus 2.83 at 70 keV and 3.67 with DCE-MRI. The difference was significant versus both 70 keV (p = 0.002) and DCE-MRI (p = 0.041).</p><p><br />
Lesion detection was numerically highest with 40-keV VMI (90.9%), compared with 77.3% at 70 keV and 84.8% with DCE-MRI, while lesion CNR was substantially higher at 40 keV (8.95 vs 5.78, p &lt; 0.001). Pairwise differences in detection rate, however, were not statistically significant.</p><p><br />
The gain in conspicuity came with trade-offs: 40-keV images had more perceived noise and showed greater underestimation of lesion extent than DCE-MRI, with median directional size error −29.2% vs −15.4% (p = 0.009). The authors therefore position low-keV PCCT as a useful complementary option for lesion localization within staging CT, while MRI remains preferable for comprehensive assessment of disease extent. </p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Flink.springer.com%2Farticle%2F10.1007%2Fs11604-026-01979-7" rel="nofollow noopener ugc" role="button">Access publication</a></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/5XORRTY07FVP\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:2466078,&quot;width&quot;:2031,&quot;height&quot;:1150,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F5XORRTY07FVP%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1863,&quot;dateInserted&quot;:&quot;2026-10-13T13:18:44+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F5XORRTY07FVP%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/5XORRTY07FVP/image.png" alt="image.png" height="1150" width="2031" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/5XORRTY07FVP/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/5XORRTY07FVP/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/5XORRTY07FVP/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/5XORRTY07FVP/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/5XORRTY07FVP/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/5XORRTY07FVP/image.png 2000w, https://us.v-cdn.net/6038601/uploads/5XORRTY07FVP/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 1.</strong> Prone positioning protocol for CT acquisition. a Patient positioned prone on the CT table using the standardized support setup. b Support arrangement and intended positioning. The face is placed within the head support (yellow circle), the arms are positioned on the arm supports (green circles), and the breasts are allowed to hang freely in a dependent, non-compressive configuration in the designated opening (red rectangle) to minimize deformation.</p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Whole-body staging with local breast assessment</p></td></tr><tr><td><p><strong>Patient position</strong></p></td><td><p>Prone</p></td></tr><tr><td><p><strong>Breast positioning</strong></p></td><td><p>Dependent, non-compressive positioning using standardized support</p></td></tr><tr><td><p><strong>Acquisition phases</strong></p></td><td><p>Noncontrast, arterial, equilibrium</p></td></tr><tr><td><p><strong>Spectral reconstruction</strong></p></td><td><p>40- and 70-keV VMI</p></td></tr><tr><td><p><strong>Reconstruction slice thickness [mm]</strong></p></td><td><p>1.0</p></td></tr><tr><td><p><strong>Reconstruction increment [mm]</strong></p></td><td><p>1.0</p></td></tr><tr><td><p><strong>Matrix</strong></p></td><td><p>512 × 512</p></td></tr><tr><td><p><strong>Kernel</strong></p></td><td><p>Qr40</p></td></tr><tr><td><p><strong>Iterative reconstruction</strong></p></td><td><p>QIR 2</p></td></tr><tr><td><p><strong>Post-processing</strong></p></td><td><p>syngo.via</p></td></tr><tr><td><p><strong>Contrast concentration</strong></p></td><td><p>300 mg I/mL</p></td></tr><tr><td><p><strong>Contrast volume</strong></p></td><td><p>2 mL/kg</p></td></tr><tr><td><p><strong>Injection duration</strong></p></td><td><p>30 s</p></td></tr><tr><td><p><strong>Arterial phase</strong></p></td><td><p>50 s after contrast injection</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p><strong>22 women</strong>, median age <strong>55.5 years</strong></p></td></tr><tr><td><p><strong>Lesion conspicuity</strong></p></td><td><p>40 keV <strong>4.33</strong> vs 70 keV <strong>2.83</strong> vs DCE-MRI <strong>3.67</strong>; 40 keV vs MRI <em>p</em> = 0.041</p></td></tr><tr><td><p><strong>Detection rate</strong></p></td><td><p><strong>90.9%</strong> at 40 keV vs <strong>77.3%</strong> at 70 keV vs <strong>84.8%</strong> with DCE-MRI</p></td></tr><tr><td><p><strong>Lesion CNR</strong></p></td><td><p><strong>8.95</strong> at 40 keV vs <strong>5.78</strong> at 70 keV, <em>p</em> &lt; 0.001</p></td></tr><tr><td><p><strong>Directional size error</strong></p></td><td><p>40 keV <strong>−29.2%</strong> vs DCE-MRI <strong>−15.4%</strong>, <em>p</em> = 0.009</p></td></tr><tr><td><p><strong>Reading time</strong></p></td><td><p>40 keV <strong>174.3 s</strong> vs DCE-MRI <strong>128.8 s</strong>, <em>p</em> = 0.001</p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Kawasaki Y, Nakajima M, Yokoyama K, Yamaga E, Kato M, Katsuta L, Yamagiwa K, Hada H, Hiraishi T, Yamada H, Tsuchiya J, Kirimura S, Onishi I, Aruga T, Tateishi U. <em>Lesion conspicuity assessment of prone 40-keV virtual monoenergetic imaging in arterial-phase photon-counting CT for breast cancer: a retrospective comparison with DCE-MRI.</em> <strong>Japanese Journal of Radiology.</strong> 2026;44:1177–1188. doi:10.1007/s11604-026-01979-7.</p>]]>
        </description>
    </item>
    <item>
        <title>(2025) Braaksma | Pediatric abdomen PCCT protocol development</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/604/2025-braaksma-pediatric-abdomen-pcct-protocol-development</link>
        <pubDate>Sat, 24 Oct 2026 13:00:19 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">604@/de/discussions</guid>
        <description><![CDATA[<div><div><p>For routine pediatric abdomen PCCT, start with Quantum mode at 70 kV and low-keV VMI rather than defaulting to QuantumPlus. Based on their phantom work, the authors favor the lowest tube voltage as the ALARA-oriented starting point, and for their standard protocol they selected 70 kV Quantum mode with VMI 53 keV.</p><p><br />
Optimize image quality first through reconstruction settings before increasing acquisition dose. The authors recommend first adjusting QIR strength, kernel choice, and windowing if noise is too high, and only then considering higher IQ level, DAC changes, or QuantumPlus mode.</p><p><br />
Use high-pitch FLASH selectively, not routinely. The authors found FLASH shortened scan time but slightly increased noise and dose, so they recommend it mainly for children who cannot hold their breath long enough for a standard scan.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Flink.springer.com%2Farticle%2F10.1007%2Fs00247-025-06274-7" rel="nofollow noopener ugc" role="button">Access publication</a></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/2T3J4MAENTLK\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:884483,&quot;width&quot;:2032,&quot;height&quot;:530,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F2T3J4MAENTLK%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1288,&quot;dateInserted&quot;:&quot;2026-04-10T14:10:53+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:&quot;embed&quot;,&quot;foreignID&quot;:&quot;14&quot;,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F2T3J4MAENTLK%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/2T3J4MAENTLK/image.png" alt="image.png" height="530" width="2032" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/2T3J4MAENTLK/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/2T3J4MAENTLK/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/2T3J4MAENTLK/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/2T3J4MAENTLK/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/2T3J4MAENTLK/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/2T3J4MAENTLK/image.png 2000w, https://us.v-cdn.net/6038601/uploads/2T3J4MAENTLK/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 5. </strong>Examples of the clinical implementation of the pediatric abdomen protocol. Axial computed tomography slices in the portal venous phase after intravenous contrast demonstrate the appearances of contrast and noise of the protocol at different ages. Protocol: 70 kV Quantum mode at 53 keV virtual monoenergetic image, IQ level 120, body reconstruction kernel 40, quantum iterative reconstruction 2, 1.0-mm slice thickness. a Six-month-old girl after liver transplantation. b Two-year-old girl prior to liver transplantation. c Six-year-old boy after liver transplantation</p><p><em>Source note for image: Braaksma et al.,</em> Pediatric Radiology, <em>2025.</em></p><p></p><h2 data-id="standard-protocol"><strong>Standard Protocol</strong></h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p>Parameter</p></th><td><p></p></td></tr><tr><th><p>Scanner</p></th><td><p>NAEOTOM Alpha</p></td></tr><tr><th><p>Scan mode</p></th><td><p>Quantum</p></td></tr><tr><th><p>Siemens protocol</p></th><td><p>RoutineFlashSpiralChildAbdomenQuantum</p></td></tr><tr><th><p>Tube voltage, kV</p></th><td><p>70/90</p></td></tr><tr><th><p>Number of slices, mm</p></th><td><p>144 × 0.4</p></td></tr><tr><th><p>Z-coverage, mm</p></th><td><p>58</p></td></tr><tr><th><p>Rotation time, s</p></th><td><p>0.25</p></td></tr><tr><th><p>IQ level</p></th><td><p>120</p></td></tr><tr><th><p>Pitch</p></th><td><p>3.2 (FLASH) or 1.5</p></td></tr><tr><th><p>Care keV</p></th><td><p>On</p></td></tr><tr><th><p>CARE Dose4D</p></th><td><p>On</p></td></tr><tr><th><p>Dose adaption curve</p></th><td><p>Average</p></td></tr><tr><th><p>Tissue of interest setting</p></th><td><p>Soft tissue with contrast</p></td></tr></table></div><h4 data-id="n-a"><strong></strong></h4><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p>Reconstruction parameters</p></th><td><p></p></td></tr><tr><th><p>Reconstructed slice thickness, mm</p></th><td><p>1.0 and 3.0 (0.6 and 2.0 increment)</p></td></tr><tr><th><p>Matrix</p></th><td><p>Auto</p></td></tr><tr><th><p>VMI, keV</p></th><td><p>53</p></td></tr><tr><th><p>Kernels</p></th><td><p>Br40 (1 mm), Br40 (3 mm)</p></td></tr><tr><th><p>QIR</p></th><td><p>2</p></td></tr><tr><th><p>Window level / window width</p></th><td><p>55 / 335</p></td></tr><tr><th><p>Spectral analysis</p></th><td><p>No</p></td></tr><tr><th><p></p><p><strong>Intravenous iodine contrast</strong></p></th><td><p></p></td></tr><tr><th><p>Contrast phase</p></th><td><p>Portal venous, delay is age dependent</p></td></tr><tr><th><p>Volume and flow rate</p></th><td><p>Age and kV dependent</p></td></tr></table></div><p></p><h2 data-id="complex-protocol-not-used-clinically-optimized-for-noise-thinner-slices"><strong>Complex Protocol (not used clinically, optimized for noise &amp; thinner slices)</strong></h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p>Parameter</p></th><td><p><strong>Value</strong></p></td></tr><tr><th><p>Scanner</p></th><td><p>NAEOTOM Alpha</p></td></tr><tr><th><p>Scan mode</p></th><td><p>QuantumPlus</p></td></tr><tr><th><p>Siemens protocol</p></th><td><p>RoutineFlashSpiralChildAbdomenQuantumplus</p></td></tr><tr><th><p>Tube voltage, kV</p></th><td><p>120/140</p></td></tr><tr><th><p>Number of slices, mm</p></th><td><p>144 × 0.4</p></td></tr><tr><th><p>Z-coverage, mm</p></th><td><p>58</p></td></tr><tr><th><p>Rotation time, s</p></th><td><p>0.25</p></td></tr><tr><th><p>IQ level</p></th><td><p>120</p></td></tr><tr><th><p>Pitch</p></th><td><p>3.2 (FLASH) or 1.5</p></td></tr><tr><th><p>Care keV</p></th><td><p>On</p></td></tr><tr><th><p>CARE Dose4D</p></th><td><p>On</p></td></tr><tr><th><p>Dose adaption curve</p></th><td><p>Average</p></td></tr><tr><th><p>Tissue of interest setting</p></th><td><p>Soft tissue with contrast</p></td></tr></table></div><h3 data-id="n-a-1"><strong></strong></h3><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p>Reconstruction parameters</p></th><td><p></p></td></tr><tr><th><p>Reconstructed slice thickness, mm</p></th><td><p>0.4, 1.0 and 3.0 (0.3, 0.6 and 2.0 increment)</p></td></tr><tr><th><p>Matrix</p></th><td><p>Auto</p></td></tr><tr><th><p>VMI, keV</p></th><td><p>53 (T3D for non-contrast if needed)</p></td></tr><tr><th><p>Kernels</p></th><td><p>Br40 (1 mm), Br40 (3 mm), Bv60 (0.4, 1 mm)</p></td></tr><tr><th><p>QIR</p></th><td><p>2</p></td></tr><tr><th><p>Window level / window width</p></th><td><p>55 / 335</p></td></tr><tr><th><p>Spectral analysis</p></th><td><p>Iodine map, virtual non-contrast</p></td></tr><tr><th><p><strong>Intravenous iodine contrast</strong></p></th><td><p></p></td></tr><tr><th><p>Contrast phase</p></th><td><p>Portal venous, delay is age dependent</p></td></tr><tr><th><p>Volume and flow rate</p></th><td><p>Age and kV dependent</p></td></tr></table></div><p><br />
Source: Braaksma J, Greuter MJW, Verhagen MV. Physics and rationale in pediatric abdomen photon counting detector computed tomography: an investigative review towards development of a pediatric abdomen protocol. <em>Pediatr Radiol.</em> 2025;55:1556–1567. doi:10.1007/s00247-025-06274-7</p>]]>
        </description>
    </item>
    <item>
        <title>(2026) Stammen | Contrast media reduction for CTPA using PCCT vs EID-CT</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/603/2026-stammen-contrast-media-reduction-for-ctpa-using-pcct-vs-eid-ct</link>
        <pubDate>Sat, 24 Oct 2026 12:48:00 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">603@/de/discussions</guid>
        <description><![CDATA[<div><div><p>PCCT reduced iodine load by 26.7% and CTDIvol by 24.4% compared to highly individualized EID-CT protocols, confirming real-world dose efficiency even against optimized reference standards. Diagnostic image quality (IQ) remained equivalent, with no significant differences in attenuation or SNR and only a minor, clinically irrelevant CNR reduction proximally.</p><p></p><p>A TBW-adapted CM protocol with fixed 120 kV and 55 keV VMI is sufficient on PCCT, simplifying workflow while enabling contrast and radiation dose savings.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Flink.springer.com%2Farticle%2F10.1007%2Fs00330-025-12054-6" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/PO5WSECN3UUG\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:260192,&quot;width&quot;:685,&quot;height&quot;:385,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FPO5WSECN3UUG%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1978,&quot;dateInserted&quot;:&quot;2026-10-24T12:46:49+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FPO5WSECN3UUG%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/PO5WSECN3UUG/image.png" alt="image.png" height="385" width="685" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/PO5WSECN3UUG/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/PO5WSECN3UUG/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/PO5WSECN3UUG/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/PO5WSECN3UUG/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/PO5WSECN3UUG/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/PO5WSECN3UUG/image.png 2000w, https://us.v-cdn.net/6038601/uploads/PO5WSECN3UUG/image.png" sizes="100vw" /></a>
    </span>
</span>
<span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/AA9A3CCBDRAP\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:1295656,&quot;width&quot;:1350,&quot;height&quot;:1463,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FAA9A3CCBDRAP%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1979,&quot;dateInserted&quot;:&quot;2026-10-24T12:46:57+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FAA9A3CCBDRAP%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/AA9A3CCBDRAP/image.png" alt="image.png" height="1463" width="1350" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/AA9A3CCBDRAP/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/AA9A3CCBDRAP/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/AA9A3CCBDRAP/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/AA9A3CCBDRAP/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/AA9A3CCBDRAP/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/AA9A3CCBDRAP/image.png 2000w, https://us.v-cdn.net/6038601/uploads/AA9A3CCBDRAP/image.png" sizes="100vw" /></a>
    </span>
</span>
<h3 data-id="figure-3">Figure 3</h3><p>Representative CT pulmonary artery scans from both a PCD-CT and an EID-CT. Images A and B were rated by both observers as moderate, and C and D were rated as good by both observers. Images E and F are examples of PE findings for both scanners.</p><h3 data-id="protocol">Protocol</h3><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p>Detail</p></th><th><p>Protocol</p></th></tr><tr><td><p>Scanner</p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p>Scan area</p></td><td><p>Pulmonary arteries (CTPA)</p></td></tr><tr><td><p>Scan mode</p></td><td><p>QuantumPlus </p></td></tr><tr><td><p>Tube voltage [kV]</p></td><td><p>120</p></td></tr><tr><td><p>IQ level</p></td><td><p>70 (CARE keV)</p></td></tr><tr><td><p>Dose modulation</p></td><td><p>CARE Dose4D, CARE keV</p></td></tr><tr><td><p>CTDIvol [mGy]</p></td><td><p>3.4 +/- 1.1</p></td></tr><tr><td><p>DLP [mGy*cm]</p></td><td><p>n.r.</p></td></tr><tr><td><p>Effective dose [mSv]</p></td><td><p>n.r.</p></td></tr><tr><td><p>Rotation time [s]</p></td><td><p>0.25</p></td></tr><tr><td><p>Pitch</p></td><td><p>2.4</p></td></tr><tr><td><p>Slice collimation [mm]</p></td><td><p>144 × 0.4</p></td></tr><tr><td><p>Slice width [mm]</p></td><td><p>1.0</p></td></tr><tr><td><p>Reconstruction increment [mm]</p></td><td><p>0.8</p></td></tr><tr><td><p>Reconstruction kernel</p></td><td><p>Bv40</p></td></tr><tr><td><p>Iterative reconstruction</p></td><td><p>QIR (strength 3)</p></td></tr><tr><td><p>keV level [keV]</p></td><td><p>55 (VMI)</p></td></tr><tr><td><p>Contrast agent</p></td><td><p>Ultravist 300 / 370</p></td></tr><tr><td><p>Contrast volume</p></td><td><p>TBW-adapted (≈10.2 +/- 1.4 g iodine total load)</p></td></tr><tr><td><p>Flow rate</p></td><td><p>n.r. (fixed injection duration around 8 s)</p></td></tr><tr><td><p>Saline chaser</p></td><td><p>40 mL</p></td></tr><tr><td><p>Acquisition phase</p></td><td><p>Pulmonary arterial (test bolus + peak + 7 s)</p></td></tr></table></div><p>Stammen L, Hoeijmakers EJI, Flohr TG, et al. <em>Reducing contrast media dosage for pulmonary embolism CTPA in PCCT</em>. Eur Radiol. 2026;36:3206–3215. doi:10.1007/s00330-025-12054-6</p>]]>
        </description>
    </item>
    <item>
        <title>(2026) Gkizas | PCCT-derived ECV quantifies myocardial fibrosis in HCM</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/602/2026-gkizas-pcct-derived-ecv-quantifies-myocardial-fibrosis-in-hcm</link>
        <pubDate>Sat, 24 Oct 2026 07:57:00 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">602@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In 30 patients with genotype-confirmed hypertrophic cardiomyopathy, global PCCT-derived extracellular volume (ECV) closely matched cardiac MRI: 30.0 ± 4.8% vs 30.62 ± 4.2% (p = 0.59), with a small Bland–Altman bias of 0.6% and limits of agreement of −1.49% to 2.75%.</p><p></p><p>Agreement was also strong regionally, with segmental correlations of r = 0.89 at the base and r = 0.85 at both mid-ventricular and apical levels. Interobserver reproducibility was excellent for PCCT-derived ECV (ICC 0.982) and MRI-derived ECV (ICC 0.921).   </p><p></p><p>A global PCCT ECV threshold of 33.4% identified patients with LGE ≥15% with 80% sensitivity and 76% specificity (AUC 0.77). PCCT ECV also correlated with functional and biomarker severity, including peak VO₂ (r = −0.76) and NT-proBNP (r = 0.59), supporting its potential as a combined coronary and myocardial tissue-characterization biomarker. </p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fwww.sciencedirect.com%2Fscience%2Farticle%2Fpii%2FS2211568425001640" rel="nofollow noopener ugc" role="button">Access publication</a></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/MX9WJ9HVMLKQ\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:591950,&quot;width&quot;:1000,&quot;height&quot;:575,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FMX9WJ9HVMLKQ%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1976,&quot;dateInserted&quot;:&quot;2026-10-24T07:56:12+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FMX9WJ9HVMLKQ%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/MX9WJ9HVMLKQ/image.png" alt="image.png" height="575" width="1000" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/MX9WJ9HVMLKQ/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/MX9WJ9HVMLKQ/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/MX9WJ9HVMLKQ/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/MX9WJ9HVMLKQ/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/MX9WJ9HVMLKQ/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/MX9WJ9HVMLKQ/image.png 2000w, https://us.v-cdn.net/6038601/uploads/MX9WJ9HVMLKQ/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Fig. 4. </strong>47-year-old man with hypertrophic cardiomyopathy due to MYBPC3 mutation (hematocrit level, 0.46 %).</p><p>(A) Short-axis view of late iodine enhancement (iodine map) with photon-counting CT.<br />
(B) Short-axis view of late gadolinium enhancement with cardiac MRI.<br />
(C-D) Extracellular volume polar maps (17-segment model of the American Heart Association) derived from C. Photon-counting CT and D. Cardiac MRI.    Eingefügter Text(20261024-07534…</p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Cardiac</p></td></tr><tr><td><p><strong>CCTA mode</strong></p></td><td><p>High-pitch or sequential, prospectively ECG-triggered depending on heart rate and rhythm</p></td></tr><tr><td><p><strong>tube voltage [kV]</strong></p></td><td><p>90</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>56</p></td></tr><tr><td><p><strong>reconstruction slice thickness [mm]</strong></p></td><td><p>0.4</p></td></tr><tr><td><p><strong> reconstruction increment [mm]</strong></p></td><td><p>0.4</p></td></tr><tr><td><p><strong>kernel</strong></p></td><td><p>Bv56</p></td></tr><tr><td><p><strong>iterative reconstruction</strong></p></td><td><p>QIR 4</p></td></tr><tr><td><p><strong>Late iodine enhancement mode</strong></p></td><td><p>High-pitch spectral acquisition</p></td></tr><tr><td><p><strong>Late iodine enhancement timing</strong></p></td><td><p>5 min after reinjection</p></td></tr><tr><td><p><strong>Late iodine enhancement tube voltage [kV]</strong></p></td><td><p>120</p></td></tr><tr><td><p><strong>Late iodine enhancement IQ level</strong></p></td><td><p>100</p></td></tr><tr><td><p><strong>ECG gating / phase</strong></p></td><td><p>ECG-triggered, fixed absolute RR interval 280 ms</p></td></tr><tr><td><p><strong>VMI</strong></p></td><td><p>40 keV</p></td></tr><tr><td><p><strong>LIE reconstruction slice thickness [mm]</strong></p></td><td><p>2</p></td></tr><tr><td><p><strong>LIE reconstruction increment [mm]</strong></p></td><td><p>1</p></td></tr><tr><td><p><strong>LIE kernel</strong></p></td><td><p>Qr44</p></td></tr><tr><td><p><strong>LIE iterative reconstruction</strong></p></td><td><p>QIR 4</p></td></tr><tr><td><p><strong>Spectral reconstruction</strong></p></td><td><p>Iodine maps</p></td></tr><tr><td><p><strong>Contrast</strong></p></td><td><p>Iobitridol 350 mg I/mL</p></td></tr><tr><td><p><strong>CCTA contrast volume</strong></p></td><td><p>50 mL</p></td></tr><tr><td><p><strong>LIE contrast volume</strong></p></td><td><p>Additional 50 mL</p></td></tr><tr><td><p><strong>Injection rate</strong></p></td><td><p>4–5 mL/s</p></td></tr><tr><td><p><strong>Saline flush</strong></p></td><td><p>20 mL</p></td></tr><tr><td><p><strong>Post-processing</strong></p></td><td><p>syngo.via VB40A; prototype Frontier Cardiac Functional Analyses for ECV</p></td></tr><tr><td><p><strong>Late-scan DLP [mGy·cm]</strong></p></td><td><p>106 ± 52</p></td></tr><tr><td><p><strong>Late-scan effective dose [mSv]</strong></p></td><td><p>1.47 ± 0.6</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p><strong>30 patients</strong>, 22 men; mean age ~59 years</p></td></tr><tr><td><p><strong>Global ECV</strong></p></td><td><p>PCCT <strong>30.0 ± 4.8%</strong> vs MRI <strong>30.62 ± 4.2%</strong>, <em>p</em> = 0.59</p></td></tr><tr><td><p><strong>PCCT–MRI agreement</strong></p></td><td><p>Bias <strong>0.6%</strong>, limits of agreement <strong>−1.49% to 2.75%</strong></p></td></tr><tr><td><p><strong>Segmental correlation</strong></p></td><td><p>Basal <strong>r = 0.89</strong>; mid-ventricular <strong>r = 0.85</strong>; apical <strong>r = 0.85</strong></p></td></tr><tr><td><p><strong>ECV threshold for LGE ≥15%</strong></p></td><td><p><strong>33.4%</strong>; sensitivity <strong>80%</strong>, specificity <strong>76%</strong>, AUC <strong>0.77</strong></p></td></tr><tr><td><p><strong>Clinical correlations</strong></p></td><td><p>Peak VO₂ <strong>r = −0.76</strong>; NT-proBNP <strong>r = 0.59</strong></p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Gkizas C, Longere B, Bechrouri S, Ridon H, Rodriguez Musso A, Haidar M, Croisille C, Montaigne D, De Groote P, Pontana F. <em>Photon-counting CT myocardial extracellular volume: A non-invasive biomarker for fibrosis in patients with hypertrophic cardiomyopathy.</em> <strong>Diagnostic and Interventional Imaging.</strong> 2026;107(1):17–24. doi:10.1016/j.diii.2025.09.001. </p>]]>
        </description>
    </item>
    <item>
        <title>PCCT for coronary assessment and functional imaging in Takotsubo cardiomyopathy</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/511/pcct-for-coronary-assessment-and-functional-imaging-in-takotsubo-cardiomyopathy</link>
        <pubDate>Thu, 01 Oct 2026 10:02:30 +0000</pubDate>
        <category>Case reports</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">511@/de/discussions</guid>
        <description><![CDATA[<h3 data-id="balint-szilveszter-borbala-vattay-peter-csongor-milan-vecsey-nagy"><a data-username="Bálint Szilveszter" data-userid="159" rel="nofollow" href="https://connect.siemens-healthineers.com/de/profile/159/B%C3%A1lint%20Szilveszter">@Bálint Szilveszter</a>, Borbála Vattay, Péter Csongor,  <a data-username="Milan Vecsey-Nagy" data-userid="755" rel="nofollow" href="https://connect.siemens-healthineers.com/de/profile/755/Milan%20Vecsey-Nagy">@Milan Vecsey-Nagy</a><br /></h3><h4 data-id="varosmajor-heart-and-vascular-centre-cardiac-ct-working-group-budapest-hungary">Városmajor Heart and Vascular Centre, Cardiac CT Working Group, Budapest, Hungary</h4><h4 data-id="september-29-2026">September 29, 2026</h4><p></p><div data-embedjson="{&quot;height&quot;:&quot;360&quot;,&quot;width&quot;:&quot;640&quot;,&quot;isKnowledge&quot;:false,&quot;url&quot;:&quot;https:\/\/siemens-healthineers.bynder.com\/asset\/a85e1bdc-bf2b-4a4d-94a0-dc890f0a7651\/CT-NAEOTOM-CONNECT-CaseReport-Takutsobu-2-VIDEO-2026.mp4&quot;,&quot;embedType&quot;:&quot;iframe&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fsiemens-healthineers.bynder.com%2Fasset%2Fa85e1bdc-bf2b-4a4d-94a0-dc890f0a7651%2FCT-NAEOTOM-CONNECT-CaseReport-Takutsobu-2-VIDEO-2026.mp4" rel="nofollow noopener ugc">
        https://siemens-healthineers.bynder.com/asset/a85e1bdc-bf2b-4a4d-94a0-dc890f0a7651/CT-NAEOTOM-CONNECT-CaseReport-Takutsobu-2-VIDEO-2026.mp4
    </a>
</div><div data-embedjson="{&quot;height&quot;:&quot;360&quot;,&quot;width&quot;:&quot;640&quot;,&quot;isKnowledge&quot;:false,&quot;url&quot;:&quot;https:\/\/siemens-healthineers.bynder.com\/asset\/377c2032-a4c9-4561-9e4f-d6479c6dbd1d\/CT-NAEOTOM-CONNECT-CaseReport-Takutsobu-1-VIDEO-2026.mp4&quot;,&quot;embedType&quot;:&quot;iframe&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fsiemens-healthineers.bynder.com%2Fasset%2F377c2032-a4c9-4561-9e4f-d6479c6dbd1d%2FCT-NAEOTOM-CONNECT-CaseReport-Takutsobu-1-VIDEO-2026.mp4" rel="nofollow noopener ugc">
        https://siemens-healthineers.bynder.com/asset/377c2032-a4c9-4561-9e4f-d6479c6dbd1d/CT-NAEOTOM-CONNECT-CaseReport-Takutsobu-1-VIDEO-2026.mp4
    </a>
</div><p></p><h2 data-id="history">History</h2><h4 data-id="why-did-the-patient-come-for-a-ct-exam">Why did the patient come for a CT exam?</h4><p>A 63-year-old female patient, presenting with typical chest pain, was referred for coronary imaging to clarify her coronary status and exclude acute coronary syndrome.</p><p>Laboratory tests showed marked myocardial injury and neurohormonal activation, with highly elevated high-sensitivity troponin I of 2032 ng/L, CK of 565 U/L, CK-MB of 43 U/L, and NT-proBNP of 1589.7 pg/mL.</p><p>Mild inflammatory activity was also present, with leukocytosis of 12.86 G/L, neutrophilia, and mildly elevated CRP of 6.2 mg/L. Renal function was preserved, with an eGFR of 107 mL/min.</p><p>An initial transthoracic echocardiography showed mildly to moderately reduced global left ventricular systolic function, with a left ventricular ejection fraction of 42% by Simpson’s method. Regional wall motion abnormalities were present, with hypokinesis of the midventricular segments and near akinesia of the apical third of the left ventricle. Mild mitral regurgitation and moderate tricuspid regurgitation were present.</p><p>Right ventricular size was normal, and right ventricular longitudinal systolic function was preserved. The estimated pulmonary artery systolic pressure was mildly elevated.</p><p>Coronary CT angiography (CCTA) was recommended to exclude obstructive coronary artery disease.</p><p></p><h2 data-id="diagnosis">Diagnosis</h2><h4 data-id="what-were-the-ct-findings-and-what-happened-to-the-patient-after-ct">What were the CT findings, and what happened to the patient after CT?</h4><p>Coronary CT angiography was performed using a dual-source photoncounting detector CT scanner (Siemens NAEOTOM Alpha Peak).</p><p>The estimated CCTA radiation dose was 397 mGy·cm. Iomeron 400 was administered with a contrast volume of 90 mL and a flow of 5.5 ml/s. The patient received 0.8 mg sublingual nitroglycerin prior to the examination. Heart rate during the scan was 93–97/min with sinus rhythm.</p><p>The coronary calcium score was 4.</p><p>The patient had a left-dominant coronary artery system and normal origin and course of the coronary arteries. No relevant coronary atherosclerosis or luminal stenosis were detected. Only minimal vessel wall irregularities were observed, with small specks of calcified plaque in the distal left anterior descending (LAD) artery (Figure 1. and 2.)</p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/F7ZVE7HEDSFS\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:1518105,&quot;width&quot;:2000,&quot;height&quot;:1150,&quot;displaySize&quot;:&quot;medium&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FF7ZVE7HEDSFS%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1716,&quot;dateInserted&quot;:&quot;2026-09-03T08:50:20+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FF7ZVE7HEDSFS%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/F7ZVE7HEDSFS/image.png" alt="image.png" height="1150" width="2000" data-display-size="medium" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/F7ZVE7HEDSFS/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/F7ZVE7HEDSFS/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/F7ZVE7HEDSFS/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/F7ZVE7HEDSFS/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/F7ZVE7HEDSFS/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/F7ZVE7HEDSFS/image.png 2000w, https://us.v-cdn.net/6038601/uploads/F7ZVE7HEDSFS/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 1.</strong> CT coronary angiography demonstrating<br />&#13; normal coronary artery anatomy with no<br />&#13; evidence of a significant stenosis.</p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/20KUJCL6DHDD\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:40075,&quot;width&quot;:526,&quot;height&quot;:219,&quot;displaySize&quot;:&quot;medium&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F20KUJCL6DHDD%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1714,&quot;dateInserted&quot;:&quot;2026-09-03T08:34:15+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F20KUJCL6DHDD%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/20KUJCL6DHDD/image.png" alt="image.png" height="219" width="526" data-display-size="medium" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/20KUJCL6DHDD/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/20KUJCL6DHDD/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/20KUJCL6DHDD/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/20KUJCL6DHDD/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/20KUJCL6DHDD/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/20KUJCL6DHDD/image.png 2000w, https://us.v-cdn.net/6038601/uploads/20KUJCL6DHDD/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 2.</strong> Electrocardiogram (ECG) during Prospective<br />&#13; Sequential Cardiac CT Acquisition.</p><p></p><p>Coronary CT angiography was performed using a prospectively ECG-triggered sequential axial acquisition. Given the patient’s heart rate, exceeding 90 beats/min, multiphase image reconstruction was performed to ensure optimal image quality and to allow simultaneous assessment of coronary anatomy and left ventricular function.</p><p>In addition to coronary artery evaluation, the multiphase datasets were analyzed for regional wall-motion abnormalities and cardiac chamber volumetry.</p><p>Coronary analysis demonstrated non-obstructive coronary artery disease, with only minimal calcified plaque in the distal left anterior descending artery. Functional assessment revealed severe apical hypokinesis/akinesis with associated midventricular dyskinesis and preserved basal contractility, resulting in the characteristic apical ballooning pattern of a classical Takotsubo syndrome (Figure 3).</p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/BZXXKYQSKXYQ\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:169153,&quot;width&quot;:964,&quot;height&quot;:360,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FBZXXKYQSKXYQ%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1715,&quot;dateInserted&quot;:&quot;2026-09-03T08:34:58+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FBZXXKYQSKXYQ%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/BZXXKYQSKXYQ/image.png" alt="image.png" height="360" width="964" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/BZXXKYQSKXYQ/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/BZXXKYQSKXYQ/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/BZXXKYQSKXYQ/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/BZXXKYQSKXYQ/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/BZXXKYQSKXYQ/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/BZXXKYQSKXYQ/image.png 2000w, https://us.v-cdn.net/6038601/uploads/BZXXKYQSKXYQ/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 3. </strong>Left Ventricular Apical Dysfunction Consistent with Takotsubo Pattern on Multiphase Cardiac CT Functional<br />&#13; Analysis. Multiphase cardiac CT-derived left ventricular wall-motion bull’s-eye map demonstrating a regional<br />&#13; gradient of myocardial contractility with preserved basal function (red–yellow), reduced mid-ventricular motion,<br />&#13; and markedly impaired apical wall motion (blue–purple). The pattern shows an apical-dominant distribution of<br />&#13; dysfunction, consistent with a classical Takotsubo-type contraction abnormality in the appropriate clinical context<br />&#13; (left). Mildly increased left ventricular end-diastolic and end-systolic volume indices (indexed to body surface area)<br />&#13; were detected, with mildly reduced systolic function (LVEF 45%) based on the available cardiac phases.</p><p>Mildly increased left ventricular end-diastolic and end-systolic volume indices were detected, with mildly reduced systolic function (LVEF 45%).</p><p>As an incidental extracardiac finding, scattered bilateral ground-glass opacities with patchy consolidations were detected, most consistent with bronchopneumonia given the elevated inflammatory laboratory markers.</p><p></p><h2 data-id="comments">Comments</h2><h4 data-id="is-there-any-clinically-interesting-aspect-of-the-case-to-discuss-what-difference-did-photon-counting-ct-make-for-the-patient">Is there any clinically interesting aspect of the case to discuss? <br />
What difference did photon-counting CT make for the patient?</h4><p>This case highlights the clinical value of the photon-counting coronary CT angiography in the diagnostic work-up of patients presenting with acute chest pain, elevated cardiac biomarkers, and electrocardiographic changes suggestive of acute coronary syndrome.</p><p>Beyond excluding obstructive coronary artery disease, CT angiography enables comprehensive myocardial functional assessment, allowing identification of the characteristic apical ballooning pattern consistent with Takotsubo syndrome, thereby supporting a working diagnosis of MINOCA – now defined as myocardial injury with non-obstructive coronary arteries (Fifth Universal Definition of Myocardial Infarction, 2026) – with Takotsubo syndrome as the most likely final diagnosis.</p><p>The ability of photon-counting CT to simultaneously evaluate coronary anatomy, plaque burden, and left ventricular function in a single examination may facilitate rapid differentiation between obstructive coronary disease and alternative causes of acute myocardial injury, potentially reducing the need for additional imaging or invasive angiography in selected patients.</p><p>This case also demonstrates the technical advantages of photoncounting coronary CT angiography for combined coronary and functional cardiac assessment in an acute setting.</p><p>The system provides excellent spatial and temporal resolution, enabling high-quality coronary plaque evaluation, even at elevated heart rates (&gt;90 beats/min), with minimal motion artefact. This allows for a reliable assessment of the coronary lumen and detection of only minimal calcified plaque in the distal LAD.</p><p>In addition, multiphase ECG-gated reconstructions offer excellent temporal resolution for detailed motion analysis of all cardiac chambers, permitting accurate depiction of regional wall-motion abnormalities.</p><p>Overall, the combination of high spatial fidelity for plaque characterization and high temporal resolution for functional assessment underscores the robustness of photon-counting CT for integrated “one-stop” coronary and myocardial evaluation, even under challenging heart rate conditions.</p><p>Note that the role of CT in this case should be interpreted within its limits. CCTA was requested to rule out obstructive coronary artery disease in a haemodynamically stable patient; because of the elevated heart rate, a multiphase acquisition was used, which additionally depicted the wall-motion pattern. CCTA can therefore exclude obstructive coronary artery disease and further reinforce echocardiographic findings suggestive of Takotsubo syndrome, but it does not establish the diagnosis on its own. Current recommendations advise cardiac magnetic resonance imaging when the final diagnosis remains uncertain—particularly in the presence of inflammatory red flags—and follow-up echocardiography to document recovery of left ventricular function. In our patient, who was referred to our centre for CT only, further diagnostic work-up and follow-up were performed at the referring institution.</p><p></p><h2 data-id="examination-protocol">Examination Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><td><p><strong>Scanner</strong></p></td><td><p><strong>NAEOTOM Alpha.Peak</strong></p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Heart</p></td></tr><tr><td><p><strong>Scan mode</strong></p></td><td><p>QuantumPlus</p></td></tr><tr><td><p><strong>Scan length</strong></p></td><td><p>126 mm</p></td></tr><tr><td><p><strong>Scan direction</strong></p></td><td><p>Cranio-caudal</p></td></tr><tr><td><p><strong>Scan time</strong></p></td><td><p>2,38 s</p></td></tr><tr><td><p><strong>Tube voltage</strong></p></td><td><p>140 kV</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>65</p></td></tr><tr><td><p><strong>Dose modulation</strong></p></td><td><p>CARE Dose4D</p></td></tr><tr><td><p><strong>CTDIvol</strong></p></td><td><p>31.5 mGy</p></td></tr><tr><td><p><strong>DLP</strong></p></td><td><p>397 mGy*cm</p></td></tr><tr><td><p><strong>Rotation time</strong></p></td><td><p>0.25 s</p></td></tr><tr><td><p><strong>Pitch</strong></p></td><td><p>not applicable</p></td></tr><tr><td><p><strong>Slice collimation</strong></p></td><td><p>0.4 mm</p></td></tr><tr><td><p><strong>Slice width</strong></p></td><td><p>0.4 mm</p></td></tr><tr><td><p><strong>Reconstruction increment</strong></p></td><td><p>0.4 mm</p></td></tr><tr><td><p><strong>Reconstruction kernel</strong></p></td><td><p>Bv56 Q4</p></td></tr><tr><td><p><strong>keV level</strong></p></td><td><p>55</p></td></tr><tr><td><p><strong>Spectral reconstruction</strong></p></td><td><p>Monoenergetic Plus</p></td></tr><tr><td><p><strong>Heart rate</strong></p></td><td><p>95/min</p></td></tr><tr><td><p><strong>Contrast</strong></p></td><td><p>Iomeron 400, 400 mg/mL</p></td></tr><tr><td><p><strong>Volume</strong></p></td><td><p>95 mL + 10mL saline pacer + 30mL saline chaser</p></td></tr><tr><td><p><strong>Flow rate</strong></p></td><td><p>5.5 mL/s</p></td></tr><tr><td><p><strong>Start delay</strong></p></td><td><p>Bolus tracking triggered at 140 HU (CCTA)</p></td></tr></table></div><p></p><h2 data-id="disclaimer">DISCLAIMER</h2><p>The statements by customers of Siemens Healthineers described herein are based on results that were achieved in the customer’s unique setting. Because there is no “typical” hospital and many variables exist (e.g., hospital size, case mix, level of IT and/or automation adoption) there can be no guarantee that other customers will achieve the same results.</p><p>The products/features (mentioned herein) are not commercially available in all countries. Their future availability cannot be guaranteed.</p>]]>
        </description>
    </item>
    <item>
        <title>PCCT for myocardial tissue characterization in acute myocarditis: comparison with CMR</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/512/pcct-for-myocardial-tissue-characterization-in-acute-myocarditis-comparison-with-cmr</link>
        <pubDate>Mon, 28 Sep 2026 20:07:05 +0000</pubDate>
        <category>Case reports</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">512@/de/discussions</guid>
        <description><![CDATA[<p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/0NA911YUF6CE\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:285496,&quot;width&quot;:581,&quot;height&quot;:339,&quot;displaySize&quot;:&quot;inline&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F0NA911YUF6CE%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1720,&quot;dateInserted&quot;:&quot;2026-09-03T08:59:40+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_inline&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F0NA911YUF6CE%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/0NA911YUF6CE/image.png" alt="image.png" height="339" width="581" data-display-size="inline" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/0NA911YUF6CE/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/0NA911YUF6CE/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/0NA911YUF6CE/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/0NA911YUF6CE/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/0NA911YUF6CE/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/0NA911YUF6CE/image.png 2000w, https://us.v-cdn.net/6038601/uploads/0NA911YUF6CE/image.png" sizes="100vw" /></a>
    </span>
</span>
 </p><h3 data-id="ferenc-imre-suhai-barbara-sipos-milan-vecsey-nagy-balint-szilveszter">Ferenc Imre Suhai, Barbara Sipos, Milán Vecsey-Nagy, Bálint Szilveszter<br /></h3><h4 data-id="varosmajor-heart-and-vascular-centre-cardiac-ct-working-group-budapest-hungary">Városmajor Heart and Vascular Centre, Cardiac CT Working Group, Budapest, Hungary</h4><h4 data-id="september-29-2026">September 29, 2026</h4><p></p><div data-embedjson="{&quot;height&quot;:&quot;360&quot;,&quot;width&quot;:&quot;640&quot;,&quot;isKnowledge&quot;:false,&quot;url&quot;:&quot;https:\/\/siemens-healthineers.bynder.com\/asset\/2a67d11b-415e-4e25-a74a-42d82ff7a4d5\/CT-NAEOTOM-CONNECT-CaseReport-BalintMyocard-1-VIDEO-2026.mp4&quot;,&quot;embedType&quot;:&quot;iframe&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fsiemens-healthineers.bynder.com%2Fasset%2F2a67d11b-415e-4e25-a74a-42d82ff7a4d5%2FCT-NAEOTOM-CONNECT-CaseReport-BalintMyocard-1-VIDEO-2026.mp4" rel="nofollow noopener ugc">
        https://siemens-healthineers.bynder.com/asset/2a67d11b-415e-4e25-a74a-42d82ff7a4d5/CT-NAEOTOM-CONNECT-CaseReport-BalintMyocard-1-VIDEO-2026.mp4
    </a>
</div><div data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/3BV4MQZ2KVN7\/26-358-cr-photoncountingct-myocardial-tissue.pdf&quot;,&quot;name&quot;:&quot;26-358_CR_PhotonCountingCT_Myocardial_Tissue.pdf&quot;,&quot;type&quot;:&quot;application\/pdf&quot;,&quot;size&quot;:619666,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F3BV4MQZ2KVN7%2F26-358-cr-photoncountingct-myocardial-tissue.pdf&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1723,&quot;dateInserted&quot;:&quot;2026-09-03T09:10:09+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;file&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <a rel="nofollow" href="https://connect.siemens-healthineers.com/api/v2/media/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F3BV4MQZ2KVN7%2F26-358-cr-photoncountingct-myocardial-tissue.pdf" download="" aria-label="26-358_CR_PhotonCountingCT_Myocardial_Tissue.pdf">
        26-358_CR_PhotonCountingCT_Myocardial_Tissue.pdf
    </a>
</div>
<h2 data-id="history">History</h2><h4 data-id="why-did-the-patient-come-for-a-ct-exam">Why did the patient come for a CT exam?</h4><p>A 29-year-old male patient, with a history of hypertension and bronchial asthma, was referred to the Városmajor Heart and Vascular Centre from the emergency department due to acute retrosternal chest pain and markedly elevated cardiac biomarkers.</p><p>His symptoms occurred in the setting of a recent upper respiratory tract infection with fever and follicular tonsillitis. He had been treated with azithromycin before admission.</p><p>At the emergency department, laboratory tests showed significant inflammatory activity and myocardial injury, with elevated C-reactive protein, leukocytosis with neutrophilia, and markedly elevated high-sensitivity troponin T, which increased from 1302 ng/L to 1689 ng/L. CK-MB was also elevated.</p><p>Electrocardiography showed ST-segment elevation in the inferior leads with reciprocal ST depression in aVL.</p><p>Transthoracic echocardiography revealed preserved left ventricular systolic function without regional wall motion abnormalities. Given the combination of acute chest pain, elevated troponin, inflammatory syndrome, and ECG changes, acute myocarditis was suspected; however, acute coronary syndrome had to be excluded.</p><p>Therefore, a coronary CT angiography was performed as the initial imaging modality to assess the coronary status. Cardiac magnetic resonance imaging was subsequently performed for detailed myocardial tissue characterization and confirmation of the diagnosis.</p><p></p><h2 data-id="diagnosis">Diagnosis</h2><h4 data-id="what-were-the-ct-findings-and-what-happened-to-the-patient-after-ct">What were the CT findings, and what happened to the patient after CT?</h4><p>Coronary CT angiography was performed using a photon-counting detector CT on a Siemens NAEOTOM Alpha Peak scanner. Native calcium scoring, contrast-enhanced ECG-gated coronary CTA, and a 6-minute delayed-phase acquisition for myocardial tissue characterization and visual assessment of late iodine enhancement were performed.</p><p>The estimated radiation dose was 350 mGy·cm for the CCTA and 245 mGy·cm for the delayed scan.</p><p>Iomeron 400 was administered with a contrast volume of 80 mL. The patient received 75 mg oral metoprolol and 0.8 mg sublingual nitroglycerin prior to the examination. Heart rate during the scan was 85/min with sinus rhythm.</p><p>The coronary calcium score was 0.</p><p>The patient had a left-dominant coronary artery system and normal origin and course of the coronary arteries. No relevant coronary atherosclerosis or luminal stenosis were detected (Figure 1).</p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/EXORB42PL2G3\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:519928,&quot;width&quot;:1091,&quot;height&quot;:641,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FEXORB42PL2G3%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1717,&quot;dateInserted&quot;:&quot;2026-09-03T08:57:07+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FEXORB42PL2G3%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/EXORB42PL2G3/image.png" alt="image.png" height="641" width="1091" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/EXORB42PL2G3/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/EXORB42PL2G3/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/EXORB42PL2G3/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/EXORB42PL2G3/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/EXORB42PL2G3/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/EXORB42PL2G3/image.png 2000w, https://us.v-cdn.net/6038601/uploads/EXORB42PL2G3/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 1. </strong>CT coronary angiography demonstrating normal coronary artery anatomy with no evidence of significant stenosis.</p><p></p><p>Cardiac magnetic resonance imaging was performed on a 1.5-T Siemens Aera scanner. Short-axis images demonstrate myocardial edema on T2-weighted fat-suppressed imaging and corresponding non-ischemic late gadolinium enhancement on LGE imaging in the mid-ventricular anterior-anterolateral and inferior-inferolateral left ventricular segments, predominantly in a subepicardial to midmyocardial distribution.</p><p>The concordant presence of edema and LGE supports acute myocarditis rather than ischemic infarction.</p><p>On the 6-minute delayed-phase PCCT images, visually detectable late iodine enhancement was observed, in a non-ischemic distribution. The enhancement involved the basal and midventricular anterolateral and inferolateral regions, the inferior segment in a subepicardial to midmyocardial pattern, and the apical inferior and lateral segments (Figure 2).</p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/AMJGJSANT6IR\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:285496,&quot;width&quot;:581,&quot;height&quot;:339,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FAMJGJSANT6IR%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1718,&quot;dateInserted&quot;:&quot;2026-09-03T08:57:48+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FAMJGJSANT6IR%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/AMJGJSANT6IR/image.png" alt="image.png" height="339" width="581" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/AMJGJSANT6IR/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/AMJGJSANT6IR/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/AMJGJSANT6IR/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/AMJGJSANT6IR/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/AMJGJSANT6IR/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/AMJGJSANT6IR/image.png 2000w, https://us.v-cdn.net/6038601/uploads/AMJGJSANT6IR/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 2. </strong>Electrocardiogram (ECG) During Prospective Sequential Cardiac CT Acquisition. Multimodality comparison of an acute myocardial injury from basal to apical short-axis images. The first column shows 6-minute delayed-phase PCCT images, the second column shows corresponding color-coded LUT maps, the third column shows iodine maps, and the fourth column shows matched cardiac MRI late gadolinium enhancement images. Blue arrows indicate areas of abnormal myocardial iodine enhancement on PCCT and corresponding non-ischemic myocardial injury on CMR. The abnormalities are predominantly located in the anterolateral, inferolateral, and inferior left ventricular segments, with a subepicardial to midmyocardial distribution, consistent with acute myocarditis.</p><p></p><p>The visual delayed iodine enhancement pattern indicated abnormal myocardial tissue characteristics and supported the diagnosis of an acute myocarditis.</p><p>Cardiac magnetic resonance (CMR) imaging was performed on a 1.5-T Siemens Aera scanner using Gadovist contrast agent. The CMR protocol included cine b-SSFP imaging, T2-weighted imaging, native T1 and T2 mapping, first-pass perfusion, late gadolinium enhancement imaging, and post-contrast T1 mapping for extracellular volume assessment.</p><p>CMR demonstrated mildly reduced left ventricular systolic function, with an LVEF of 45.4%. Left ventricular volumes were within the normal range, stroke volume was preserved, and left ventricular myocardial mass was normal. Right ventricular size and systolic function were preserved, with an RVEF of 52.2%.</p><p>T2-weighted images demonstrated myocardial edema in the basal and midventricular anterolateral and inferolateral segments, the midventricular inferior and partly anterior segments, and the apical inferior segment, predominantly in a subepicardial to midmyocardial distribution (Figure 3).</p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/X3RIEB831IY0\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:146490,&quot;width&quot;:898,&quot;height&quot;:383,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FX3RIEB831IY0%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1719,&quot;dateInserted&quot;:&quot;2026-09-03T08:59:00+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FX3RIEB831IY0%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/X3RIEB831IY0/image.png" alt="image.png" height="383" width="898" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/X3RIEB831IY0/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/X3RIEB831IY0/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/X3RIEB831IY0/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/X3RIEB831IY0/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/X3RIEB831IY0/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/X3RIEB831IY0/image.png 2000w, https://us.v-cdn.net/6038601/uploads/X3RIEB831IY0/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 3. </strong>Cardiac MRI confirmation of acute myocarditis: myocardial edema and non-ischemic injury</p><p></p><p>After gadolinium administration, pathological late gadolinium enhancement was observed in the same edematous myocardial segments.</p><p>Parametric mapping further supported the diagnosis. Native T1, T2, and ECV values were elevated in the affected lateral myocardial regions. ECV was increased in the basal lateral wall and midventricular lateral wall, while septal values remained within the normal range.</p><p>Overall, the CMR findings confirmed acute myocarditis with mildly reduced left ventricular ejection fraction and mild diffuse hypokinesis.</p><p></p><h2 data-id="comments">Comments</h2><h4 data-id="is-there-any-clinically-interesting-aspect-of-the-case-to-discuss-what-difference-did-photon-counting-ct-make-for-the-patient">Is there any clinically interesting aspect of the case to discuss? <br />
What difference did photon-counting CT make for the patient?</h4><p>This case highlights the potential diagnostic value of photon-counting CT in patients presenting with acute chest pain, elevated cardiac biomarkers, and electrocardiographic changes suggestive of acute coronary syndrome.</p><p>The 6-minute delayed-phase acquisition allowed visual assessment of myocardial iodine retention. Delayed-phase spectral reconstruction using monoenergetic plus images and color-coded iodine maps improved visualization of myocardial iodine distribution and enabled myocardial tissue characterization.</p><p>Focal subepicardial to mid-myocardial iodine accumulation indicated abnormal myocardial tissue characteristics. The distribution of delayed iodine enhancement corresponded closely to the edema, late gadolinium enhancement, and elevated CMR-derived ECV values in the affected myocardial segments.</p><p>An important limitation is the additional radiation exposure associated with the delayed-phase of CT imaging. This is particularly relevant in young patients, in whom radiation dose should be carefully justified and minimized.</p><p>Therefore, delayed-phase PCCT acquisitions should not be considered for routine in all patients with suspected myocarditis, but rather reserved for selected cases in which the expected diagnostic benefit outweighs the additional dose, such as when CMR is contraindicated, unavailable, or unlikely to be feasible.</p><p>PCCT could serve as a comprehensive first-line imaging tool in selected patients with suspected myocarditis and possible acute coronary syndrome. Its inherent spectral capabilities, together with its high spatial resolution, allow rapid exclusion of obstructive coronary artery disease while also enabling visual myocardial tissue characterization through delayed-phase spectral imaging, monoenergetic reconstructions, and iodine mapping.</p><p>These capabilities support comprehensive coronary artery evaluation while simultaneously providing additional insights into myocardial tissue characteristics within a single examination.</p><p></p><h2 data-id="examination-protocol">Examination Protocol</h2><p></p><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha.Peak</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Heart</p></td></tr><tr><td><p><strong>Scan mode</strong></p></td><td><p>QuantumPlus + HD (combined spectral and HighRes)</p></td></tr><tr><td><p><strong>Scan length</strong></p></td><td><p>141 mm</p></td></tr><tr><td><p><strong>Scan direction</strong></p></td><td><p>Cranio-caudal</p></td></tr><tr><td><p><strong>Scan time</strong></p></td><td><p>0.507 s</p></td></tr><tr><td><p><strong>Tube voltage</strong></p></td><td><p>140 kV</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>100</p></td></tr><tr><td><p><strong>Dose modulation</strong></p></td><td><p>CARE Dose4D</p></td></tr><tr><td><p><strong>CTDIvol</strong></p></td><td><p>17.5 mGy</p></td></tr><tr><td><p><strong>DLP</strong></p></td><td><p>245 mGy*cm</p></td></tr><tr><td><p><strong>Rotation time</strong></p></td><td><p>0.25 s</p></td></tr><tr><td><p><strong>Pitch</strong></p></td><td><p>not applicable</p></td></tr><tr><td><p><strong>Slice collimation</strong></p></td><td><p>144 x 0.4 mm</p></td></tr><tr><td><p><strong>Slice width</strong></p></td><td><p>0.4 mm</p></td></tr><tr><td><p><strong>Reconstruction increment</strong></p></td><td><p>0.4 mm</p></td></tr><tr><td><p><strong>Reconstruction kernel</strong></p></td><td><p>Qr44 Q4</p></td></tr><tr><td><p><strong>keV level</strong></p></td><td><p>40</p></td></tr><tr><td><p><strong>Spectral reconstruction</strong></p></td><td><p>Monoenergetic Plus; iodine/water decomposition; iodine maps</p></td></tr><tr><td><p><strong>Heart rate</strong></p></td><td><p>57/min (delayed)</p></td></tr><tr><td><p><strong>Contrast</strong></p></td><td><p>Iomeron 400, 400 mg/mL</p></td></tr><tr><td><p><strong>Volume</strong></p></td><td><p>80 mL + 10mL saline pacer + 30mL saline chaser (for CCTA)</p></td></tr><tr><td><p><strong>Flow rate</strong></p></td><td><p>5 mL/s</p></td></tr><tr><td><p><strong>Start delay</strong></p></td><td><p>Bolus tracking triggered at 92 HU (CCTA), delayed acquisition performed 6 minutes after contrast administration.</p></td></tr></table></div><p></p><h2 data-id="disclaimer">Disclaimer</h2><p>The statements by customers of Siemens Healthineers described herein are based on results that were achieved in the customer’s unique setting. Because there is no “typical” hospital and many variables exist (e.g., hospital size, case mix, level of IT and/or automation adoption) there can be no guarantee that other customers will achieve the same results.</p><p>The products/features (mentioned herein) are not commercially available in all countries. Their future availability cannot be guaranteed.</p>]]>
        </description>
    </item>
    <item>
        <title>(2026) Pali | PCCT iodine maps quantify when post-thrombectomy staining becomes visible</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/533/2026-pali-pcct-iodine-maps-quantify-when-post-thrombectomy-staining-becomes-visible</link>
        <pubDate>Mon, 28 Sep 2026 07:42:23 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">533@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In 12 patients after mechanical thrombectomy, 54 ASPECTS regions ultimately developed infarction; 40/54 (74.1%) showed hyperattenuation on PCCT iodine maps, but only 15/40 (37.5%) were visually hyperattenuating on conventional brain-window images. Thus, iodine accumulation was frequently measurable before becoming visually apparent on routine grayscale CT.</p><p><br />
PCCT iodine-map attenuation very accurately predicted conventional brain-window visibility (AUC 0.997). An internally derived threshold of 8.33 HU provided 100% sensitivity and 96% specificity; patient-level bootstrap analysis yielded a median cutoff of 11.69 HU with a 95% interval of 8.33–12.67 HU.</p><p><br />
The relative brain-window attenuation increase also performed well (AUC 0.984; threshold 3.06%), but showed substantially greater cutoff variability. The practical implication is that direct iodine quantification may provide a more robust and reproducible marker of post-thrombectomy contrast staining than visual assessment alone, although the thresholds remain pilot estimates requiring external validation.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fwww.mdpi.com%2F2075-4418%2F16%2F18%2F2915" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/9R9BXTCBSQFF\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:3414671,&quot;width&quot;:3128,&quot;height&quot;:2379,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F9R9BXTCBSQFF%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1799,&quot;dateInserted&quot;:&quot;2026-09-28T07:39:14+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F9R9BXTCBSQFF%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/9R9BXTCBSQFF/image.png" alt="image.png" height="2379" width="3128" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/9R9BXTCBSQFF/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/9R9BXTCBSQFF/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/9R9BXTCBSQFF/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/9R9BXTCBSQFF/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/9R9BXTCBSQFF/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/9R9BXTCBSQFF/image.png 2000w, https://us.v-cdn.net/6038601/uploads/9R9BXTCBSQFF/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 1.</strong> Examples of IM hyperattenuating areas that are clearly hyperattenuating on BW images (a,b), only mildly hyperattenuating on BW images (c,d), or not clearly hyperattenuating on BW images (e,f).</p><p></p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha.Peak</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Brain</p></td></tr><tr><td><p><strong>Reconstructions</strong></p></td><td><p>Brain window, iodine map, virtual non-contrast</p></td></tr><tr><td><p><strong>Follow-up imaging</strong></p></td><td><p>Non-contrast CT approximately 24 h after PCCT</p></td></tr><tr><td><p><strong>Reference for infarction extent</strong></p></td><td><p>Follow-up CT using ASPECTS</p></td></tr><tr><td><p><strong>Hemorrhage assessment</strong></p></td><td><p>Virtual non-contrast images</p></td></tr><tr><td><p><strong>ROI analysis</strong></p></td><td><p>Three circular 10 mm² ROIs per affected ASPECTS region</p></td></tr><tr><td><p><strong>Reference ROI</strong></p></td><td><p>Matching contralateral region at the same axial level</p></td></tr><tr><td><p><strong>Quantitative metric</strong></p></td><td><p>Iodine-map attenuation and relative brain-window attenuation increase (%ΔHU)</p></td></tr><tr><td><p><strong>Threshold analysis</strong></p></td><td><p>ROC analysis with patient-level clustered bootstrap resampling, 2000 iterations</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p><strong>12 patients</strong>, 54 ASPECTS regions with final infarction</p></td></tr><tr><td><p><strong>Iodine-map positive regions</strong></p></td><td><p><strong>40/54 (74.1%)</strong></p></td></tr><tr><td><p><strong>Visible on brain-window images</strong></p></td><td><p><strong>15/40 (37.5%)</strong></p></td></tr><tr><td><p><strong>Iodine-map threshold</strong></p></td><td><p><strong>8.33 HU</strong>; AUC <strong>0.997</strong>, sensitivity <strong>100%</strong>, specificity <strong>96.0%</strong></p></td></tr><tr><td><p><strong>Relative BW threshold</strong></p></td><td><p><strong>3.06%</strong>; AUC <strong>0.984</strong>, sensitivity <strong>100%</strong>, specificity <strong>92.0%</strong></p></td></tr><tr><td><p><strong>Inter-reader agreement</strong></p></td><td><p>κ <strong>0.81</strong> for iodine maps; κ <strong>0.86</strong> for brain-window visibility</p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Pali M-C, Mangesius S, Eisenschink CE, Deisl P, Neumann L, Knoflach M, Pechlaner R, Glodny B, Lenhart L, Gizewski ER, Grams AE. <em>From Iodine Maps to Brain Window Images: Quantitative Visibility Thresholds of Iodine Staining Using Photon-Counting CT—A Retrospective Observational Study.</em> <strong>Diagnostics.</strong> 2026;16(18):2915. doi:10.3390/diagnostics16182915.</p>]]>
        </description>
    </item>
    <item>
        <title>(2023) Turrion Gomollon | VNI PCCT may replace true noncontrast after EVAR</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/517/2023-turrion-gomollon-vni-pcct-may-replace-true-noncontrast-after-evar</link>
        <pubDate>Fri, 04 Sep 2026 06:24:59 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">517@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In <strong>110 patients undergoing EVAR surveillance</strong>, including <strong>41 endoleaks</strong>, the authors compared conventional triphasic PCCT (true noncontrast + arterial + venous) with a biphasic strategy in which <strong>virtual noniodine (VNI) images reconstructed from the venous phase replaced the true noncontrast acquisition</strong>.</p><p></p><p>Endoleak detection was essentially preserved with VNI. Sensitivity was identical for each reader (<strong>0.95 vs 0.95</strong> and <strong>0.88 vs 0.88</strong>), while specificity remained similar; interreader agreement was substantial with both TNC (<strong>α = 0.716</strong>) and VNI (<strong>α = 0.756</strong>). Subjective image quality and noise were also comparable.  </p><p></p><p>Replacing TNC with VNI could have reduced radiation exposure by approximately <strong>38%</strong>. A caveat is imperfect calcium preservation: small aneurysm calcifications were erroneously removed in <strong>15/110 patients (14%)</strong>, although this affected endoleak diagnosis in only <strong>one case</strong>, producing a false-positive interpretation.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fwww.ovid.com%2Fjnls%2Finvestigativeradiology%2Ffulltext%2F10.1097%2Frli.0000000000000993%7Ephoton-counting-detector-ct-angiography-for-endoleak" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/BSYZ2OS2QPFH\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:472432,&quot;width&quot;:1000,&quot;height&quot;:575,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FBSYZ2OS2QPFH%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1729,&quot;dateInserted&quot;:&quot;2026-09-04T06:24:37+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FBSYZ2OS2QPFH%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/BSYZ2OS2QPFH/image.png" alt="image.png" height="575" width="1000" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/BSYZ2OS2QPFH/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/BSYZ2OS2QPFH/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/BSYZ2OS2QPFH/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/BSYZ2OS2QPFH/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/BSYZ2OS2QPFH/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/BSYZ2OS2QPFH/image.png 2000w, https://us.v-cdn.net/6038601/uploads/BSYZ2OS2QPFH/image.png" sizes="100vw" /></a>
    </span>
</span>
<p>Figure 2. Axial CT images at the level of the aortic aneurysm acquired with photon-counting detector CT in a 60-year-old man referred for follow-up imaging after EVAR: (A) true noncontrast images, (B) arterial phase, (C) venous phase with an arrow pointing at the endoleak, and (D) virtual noniodine images reconstructed from venous phase images. </p><p></p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Software</strong></p></td><td><p>VA50</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Thoracoabdominal aorta / EVAR surveillance</p></td></tr><tr><td><p><strong>Scan mode</strong></p></td><td><p>QuantumPlus multienergy</p></td></tr><tr><td><p><strong>Acquisition phases</strong></p></td><td><p>True noncontrast, arterial, venous</p></td></tr><tr><td><p><strong>Tube voltage [kV]</strong></p></td><td><p>120</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>68</p></td></tr><tr><td><p><strong>Dose modulation</strong></p></td><td><p>CARE Dose4D</p></td></tr><tr><td><p><strong>CTDIvol [mGy]</strong></p></td><td><p>TNC 5.1 ± 1.5; arterial 3.74 ± 1.4; venous 4.65 ± 1.4</p></td></tr><tr><td><p><strong>DLP [mGy·cm]</strong></p></td><td><p>TNC 353.3 ± 108.1; arterial 258.9 ± 88.8; venous 320 ± 100</p></td></tr><tr><td><p><strong>Total dose</strong></p></td><td><p>CTDIvol 13.5 ± 3.9 mGy; DLP 932.2 ± 286 mGy·cm</p></td></tr><tr><td><p><strong>Rotation time [s]</strong></p></td><td><p>0.25</p></td></tr><tr><td><p><strong>Pitch</strong></p></td><td><p>1.2</p></td></tr><tr><td><p><strong>Detector slice collimation</strong></p></td><td><p>144 × 0.4 mm</p></td></tr><tr><td><p><strong>Reconstruction slice thickness / increment [mm]</strong></p></td><td><p>2.0 / 1.6</p></td></tr><tr><td><p><strong>Iterative reconstruction</strong></p></td><td><p>QIR 4</p></td></tr><tr><td><p><strong>TNC reconstruction</strong></p></td><td><p>70-keV VMI, Br36</p></td></tr><tr><td><p><strong>VNI reconstruction</strong></p></td><td><p>70 keV, PureCalcium, Qr36; reconstructed from venous phase</p></td></tr><tr><td><p><strong>Arterial reconstruction</strong></p></td><td><p>55-keV VMI</p></td></tr><tr><td><p><strong>Venous reconstruction</strong></p></td><td><p>60-keV VMI</p></td></tr><tr><td><p><strong>Contrast</strong></p></td><td><p>Iopromide, Ultravist 370 mg I/mL</p></td></tr><tr><td><p><strong>Contrast volume</strong></p></td><td><p>70 mL total</p></td></tr><tr><td><p><strong>Contrast protocol</strong></p></td><td><p>40-mL contrast bolus → 60-mL 1:1 contrast/saline mixture → 20-mL saline chaser</p></td></tr><tr><td><p><strong>Injection rate</strong></p></td><td><p>4 mL/s</p></td></tr><tr><td><p><strong>Bolus triggering</strong></p></td><td><p>Ascending aorta; 140 HU at 90 kV</p></td></tr><tr><td><p><strong>Arterial phase</strong></p></td><td><p>12 s after trigger threshold</p></td></tr><tr><td><p><strong>Venous phase</strong></p></td><td><p>70 s after trigger threshold</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p><strong>110 patients; 41 endoleaks (37%)</strong></p></td></tr><tr><td><p><strong>Reader 1 sensitivity / specificity</strong></p></td><td><p>TNC <strong>0.95 / 0.84</strong> vs VNI <strong>0.95 / 0.86</strong></p></td></tr><tr><td><p><strong>Reader 2 sensitivity / specificity</strong></p></td><td><p>TNC <strong>0.88 / 0.98</strong> vs VNI <strong>0.88 / 0.94</strong></p></td></tr><tr><td><p><strong>Interreader agreement</strong></p></td><td><p>TNC <strong>α = 0.716</strong>; VNI <strong>α = 0.756</strong></p></td></tr><tr><td><p><strong>Subjective image quality / noise</strong></p></td><td><p>Both <strong>4 [IQR 4–5]</strong>; no significant difference</p></td></tr><tr><td><p><strong>Potential dose reduction</strong></p></td><td><p><strong>38%</strong> by eliminating the TNC acquisition</p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Turrion Gomollon AM, Mergen V, Sartoretti T, Polacin M, Nakhostin D, Puippe G, Alkadhi H, Euler A. <em>Photon-Counting Detector CT Angiography for Endoleak Detection After Endovascular Aortic Repair: Triphasic CT With True Noncontrast Versus Biphasic CT With Virtual Noniodine Imaging.</em> <strong>Investigative Radiology.</strong> 2023;58(11):816–821. doi:10.1097/RLI.0000000000000993.</p>]]>
        </description>
    </item>
    <item>
        <title>(2026) Kabakus | Is nitroglycerin still necessary for CCTA in the era of PCCT?</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/514/2026-kabakus-is-nitroglycerin-still-necessary-for-ccta-in-the-era-of-pcct</link>
        <pubDate>Thu, 03 Sep 2026 11:07:46 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">514@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In 124 patients with CCTA and invasive coronary angiography (ICA) within 30 days, diagnostic performance of PCCT CCTA was similar with and without sublingual nitroglycerin. Per-patient sensitivity was 90.3% vs 90.9%, and NPV was 86.4% vs 88.9%, with no significant differences between groups.</p><p><br />
The non-nitroglycerin group was numerically better for several diagnostic metrics—specificity 80.0% vs 63.3%, PPV 83.3% vs 71.8%, and accuracy 85.7% vs 77.0%—although none reached statistical significance. Likewise, no significant differences were found across LAD, LCx, or RCA analyses, so the findings remain hypothesis-generating rather than practice-changing.</p><p><br />
Nitroglycerin increased left-main attenuation (634 ± 169 vs 546 ± 147 HU, p = 0.029) but did not improve SNR or CNR. The authors also raise the possibility that vasodilation of healthy reference segments may exaggerate percentage stenosis relative to diseased segments, potentially contributing to false-positive CCTA findings.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fwww.sciencedirect.com%2Fscience%2Farticle%2Fpii%2FS1934592526004740" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/9NJFQUXN7ZFD\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:312399,&quot;width&quot;:713,&quot;height&quot;:507,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F9NJFQUXN7ZFD%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1728,&quot;dateInserted&quot;:&quot;2026-09-03T11:06:13+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F9NJFQUXN7ZFD%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/9NJFQUXN7ZFD/image.png" alt="image.png" height="507" width="713" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/9NJFQUXN7ZFD/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/9NJFQUXN7ZFD/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/9NJFQUXN7ZFD/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/9NJFQUXN7ZFD/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/9NJFQUXN7ZFD/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/9NJFQUXN7ZFD/image.png 2000w, https://us.v-cdn.net/6038601/uploads/9NJFQUXN7ZFD/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 3. </strong>Representative case demonstrating potential nitroglycerin-related overestimation of coronary stenosis severity. A 73-year-old man presenting with chest pain underwent photon-counting CCTA following administration of 0.4 mg sublingual nitroglycerin. (A) Curved multiplanar reformatted CCTA image demonstrates a proximal coronary artery diameter of 3.5 mm, a stenotic segment diameter of 1.8 mm, and a distal reference diameter of 3.0 mm. Based on CCTA findings, the lesion was classified as moderate stenosis (CAD-RADS 3, obstructive disease). (B) Invasive coronary angiography demonstrated slightly smaller reference vessel diameters and a slightly larger residual lumen at the stenotic segment, with measurements of 3.3 mm proximally, 1.9 mm at the stenosis, and 2.7 mm distally. Consequently, the lesion was classified as mild stenosis (non-obstructive disease) by invasive coronary angiography. The discrepancy is likely attributable to nitroglycerin-induced vasodilation of adjacent healthy coronary segments, resulting in approximately 10% enlargement of the reference vessel diameter without substantial dilation of the diseased segment. This differential vasodilatory response increases the apparent percentage stenosis on CCTA and may contribute to overestimation of stenosis severity.</p><p></p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Coronary arteries / CCTA</p></td></tr><tr><td><p><strong>Acquisition mode</strong></p></td><td><p>PCCT CCTA</p></td></tr><tr><td><p><strong>Tube voltage [kV]</strong></p></td><td><p>120</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>64</p></td></tr><tr><td><p><strong>Dose modulation</strong></p></td><td><p>Automated tube current modulation</p></td></tr><tr><td><p><strong>Rotation time [s]</strong></p></td><td><p>0.25</p></td></tr><tr><td><p><strong>Detector slice collimation</strong></p></td><td><p>120 × 0.4 mm</p></td></tr><tr><td><p><strong>Prospective ECG acquisition</strong></p></td><td><p>HR &lt;75 bpm with regular rhythm; acquisition window centered at 60–80% of RR</p></td></tr><tr><td><p><strong>Retrospective ECG acquisition</strong></p></td><td><p>HR &gt;75 bpm or irregular rhythm; full radiation at 30–50% of RR, reduced to 10% during remainder</p></td></tr><tr><td><p><strong>Reconstruction slice thickness [mm]</strong></p></td><td><p>0.4</p></td></tr><tr><td><p><strong>Reconstruction increment [mm]</strong></p></td><td><p>0.3</p></td></tr><tr><td><p><strong>Matrix</strong></p></td><td><p>512 × 512</p></td></tr><tr><td><p><strong>FOV</strong></p></td><td><p>Dedicated cardiac FOV optimized for coronary assessment</p></td></tr><tr><td><p><strong>Kernel</strong></p></td><td><p>Bv44 smooth and Bv60 sharp</p></td></tr><tr><td><p><strong>Iterative reconstruction</strong></p></td><td><p>QIR 3</p></td></tr><tr><td><p><strong>Contrast</strong></p></td><td><p>Iohexol, Omnipaque 350</p></td></tr><tr><td><p><strong>Contrast volume</strong></p></td><td><p>BMI &lt;25: 50–60 mL; BMI 25–30: 60–70 mL; BMI &gt;30: 70–80 mL</p></td></tr><tr><td><p><strong>Injection rate</strong></p></td><td><p>5 mL/s</p></td></tr><tr><td><p><strong>Bolus triggering</strong></p></td><td><p>Descending aorta; monitoring begins 8 s after injection onset; 100-HU threshold</p></td></tr><tr><td><p><strong>Post-trigger delay</strong></p></td><td><p>6 s</p></td></tr><tr><td><p><strong>Mixed bolus</strong></p></td><td><p>30 mL of 50% contrast/saline at 5 mL/s</p></td></tr><tr><td><p><strong>Saline chaser</strong></p></td><td><p>30 mL at 5 mL/s</p></td></tr><tr><td><p><strong>Venous access</strong></p></td><td><p>18-gauge IV</p></td></tr><tr><td><p><strong>Heart-rate control</strong></p></td><td><p>IV metoprolol 5-mg increments every 5 min, maximum 25 mg; target HR &lt;75 bpm</p></td></tr><tr><td><p><strong>Nitroglycerin comparison</strong></p></td><td><p>0.4 mg sublingual nitroglycerin vs no nitroglycerin</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p><strong>124 patients</strong>: 61 nitroglycerin / 63 no nitroglycerin</p></td></tr><tr><td><p><strong>Sensitivity</strong></p></td><td><p><strong>90.3% vs 90.9%</strong>, <em>p</em> = 0.940</p></td></tr><tr><td><p><strong>Specificity</strong></p></td><td><p><strong>63.3% vs 80.0%</strong>, <em>p</em> = 0.160</p></td></tr><tr><td><p><strong>PPV / NPV</strong></p></td><td><p><strong>71.8% / 86.4% vs 83.3% / 88.9%</strong></p></td></tr><tr><td><p><strong>Overall accuracy</strong></p></td><td><p><strong>77.0% vs 85.7%</strong>, <em>p</em> = 0.220</p></td></tr><tr><td><p><strong>Objective image quality</strong></p></td><td><p>LM attenuation higher with nitroglycerin (<strong>634 vs 546 HU</strong>, <em>p</em> = 0.029), but <strong>SNR and CNR unchanged</strong></p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Kabakus IM, Maisuria DM, Chamberlin JH, Mencken IL, Lopez Rodriguez CG, Sagatelian MM, Burt JR, Baruah D. <em>Is nitroglycerin still necessary for coronary CT angiography in the era of photon-counting CT?</em> <strong>Journal of Cardiovascular Computed Tomography.</strong> 2026. In Press, Corrected Proof. Available online August 22, 2026. doi:10.1016/j.jcct.2026.08.007.</p>]]>
        </description>
    </item>
    <item>
        <title>(2026) Schwartz | Mixed-bolus low-keV PCCT improves CTPA robustness</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/508/2026-schwartz-mixed-bolus-low-kev-pcct-improves-ctpa-robustness</link>
        <pubDate>Wed, 02 Sep 2026 10:51:21 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">508@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In 150 patients undergoing PCCT pulmonary angiography, the investigators compared <strong>three contrast-injection</strong> strategies while reconstructing images at 40, 55, and 67 keV. Pulmonary arterial enhancement increased progressively at lower energies, with <strong>40 keV providing the strongest vascular enhancement</strong> and highest reader preference.</p><p><br />
The <strong>most robust protocols used contrast mixed 1:1 with saline and a fixed scan delay</strong>, rather than a small undiluted bolus with bolus tracking. Both the 50/50-mL and 35/35-mL contrast/saline approaches produced no examinations with pulmonary arterial enhancement below 350 HU at 40 keV, whereas the bolus-tracking cohort had three such injections.</p><p><br />
The practical message is that <strong>PCCT can make CTPA less dependent on precise bolus timing</strong>: a longer mixed bolus, fixed timing, high-pitch acquisition, and 40-keV VMI provided a robust protocol while allowing contrast volume to be reduced to 35 mL.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fwww.sciencedirect.com%2Fscience%2Farticle%2Fabs%2Fpii%2FS0720048X25006266" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/YPP5VRMOCVL1\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:955426,&quot;width&quot;:1575,&quot;height&quot;:886,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FYPP5VRMOCVL1%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1709,&quot;dateInserted&quot;:&quot;2026-09-02T10:51:15+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FYPP5VRMOCVL1%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/YPP5VRMOCVL1/image.png" alt="image.png" height="886" width="1575" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/YPP5VRMOCVL1/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/YPP5VRMOCVL1/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/YPP5VRMOCVL1/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/YPP5VRMOCVL1/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/YPP5VRMOCVL1/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/YPP5VRMOCVL1/image.png 2000w, https://us.v-cdn.net/6038601/uploads/YPP5VRMOCVL1/image.png" sizes="100vw" /></a>
    </span>
</span>
<p></p><h2 data-id="protocol">Protocol</h2><p>For a detailed view, access the full publication.</p><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>CT pulmonary angiography</p></td></tr><tr><td><p><strong>Acquisition mode</strong></p></td><td><p>high-pitch</p></td></tr><tr><td><p><strong>Tube voltage [kV]</strong></p></td><td><p>140</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>105</p></td></tr><tr><td><p><strong>Dose modulation</strong></p></td><td><p>CareDose4D</p></td></tr><tr><td><p><strong>Detector slice collimation</strong></p></td><td><p>2 × 144 × 0.4 mm</p></td></tr><tr><td><p><strong>VMI reconstructions</strong></p></td><td><p>40, 55, and 67 keV</p></td></tr><tr><td><p><strong>Cohort 1 injection</strong></p></td><td><p>50 mL contrast / 0 mL saline; bolus tracking</p></td></tr><tr><td><p><strong>Cohort 2 injection</strong></p></td><td><p>50 mL contrast + 50 mL saline; fixed scan delay</p></td></tr><tr><td><p><strong>Cohort 3 injection</strong></p></td><td><p>35 mL contrast + 35 mL saline; fixed scan delay</p></td></tr><tr><td><p><strong>Optimized contrast strategy</strong></p></td><td><p>Equal-volume contrast/saline mixture: 50/50 mL or 35/35 mL</p></td></tr><tr><td><p><strong>Injection rate</strong></p></td><td><p>4 mL/s for optimized mixed-bolus protocols</p></td></tr><tr><td><p><strong>Scan timing</strong></p></td><td><p>Acquisition started at the end of the mixed-bolus injection</p></td></tr><tr><td><p><strong>Recommended vascular reconstruction</strong></p></td><td><p>Thin-slice 40-keV VMI</p></td></tr><tr><td><p><strong>Recommended soft-tissue reconstruction</strong></p></td><td><p>Thicker 67-keV VMI</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><ul><li><strong>Population:</strong> 150 patients, 50 per protocol; mean age <strong>60.6 ± 16.6 years</strong>. </li><li>Pulmonary arterial attenuation was significantly higher with <strong>40 &gt; 55 &gt; 67 keV</strong> (<em>p</em> &lt; 0.001). </li><li>At 40 keV, <strong>3 examinations in cohort 1</strong> had attenuation &lt;350 HU, versus <strong>none in cohorts 2 or 3</strong>. </li><li>Reader-rated vascular contrast was highest at <strong>40 keV</strong>; in cohort 2, scores were <strong>4.7 ± 0.5 at 40 keV</strong>, <strong>4.5 ± 0.7 at 55 keV</strong>, and <strong>3.6 ± 1.0 at 67 keV</strong> (<em>p</em> &lt; 0.001). </li><li>The authors favored <strong>50/50-mL or 35/35-mL contrast/saline mixtures with fixed delay</strong>, concluding that this approach minimized suboptimal enhancement and repeat examinations.  </li></ul><p><strong>Source:</strong><br />
Schwartz FR, Khanungwanitkul K, Czajkowski B, Duraisamy S, Jhala K, Hammer MM, Sodickson AD. <em>Optimizing photon-counting CT pulmonary angiography for robustness, combining mixed bolus, fixed scan delay and low keV.</em> <strong>European Journal of Radiology.</strong> 2026;195:112540. doi:10.1016/j.ejrad.2025.112540.</p>]]>
        </description>
    </item>
    <item>
        <title>(2022) Lennartz | Virtual unenhanced images show limitations for adrenal adenoma characterization</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/486/2022-lennartz-virtual-unenhanced-images-show-limitations-for-adrenal-adenoma-characterization</link>
        <pubDate>Thu, 27 Aug 2026 17:27:06 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">486@/de/discussions</guid>
        <description><![CDATA[<div><div><p>PCCT-derived virtual unenhanced (VUE) images showed no significant difference in mean attenuation compared with true unenhanced (TUE) images for adrenal adenomas (9.2 ± 8.6 HU vs 7.3 ± 8.4 HU; P = .07).</p><p><br />
Despite similar mean values, individual VUE measurements showed both over- and underestimation, resulting in 10 of 36 adrenal adenomas being misclassified when the conventional 10-HU threshold was applied to VUE images.</p><p><br />
Material decomposition was tissue dependent: VUE underestimated attenuation in muscle and the portal vein while overestimating fat attenuation, indicating that further technical optimization is required before VUE can reliably replace TUE imaging for adrenal adenoma assessment.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fpubs.rsna.org%2Fdoi%2F10.1148%2Fradiol.220919" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/JQ64337ELDFY\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:1288466,&quot;width&quot;:2000,&quot;height&quot;:1150,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FJQ64337ELDFY%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1656,&quot;dateInserted&quot;:&quot;2026-08-27T17:25:46+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FJQ64337ELDFY%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/JQ64337ELDFY/image.png" alt="image.png" height="1150" width="2000" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/JQ64337ELDFY/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/JQ64337ELDFY/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/JQ64337ELDFY/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/JQ64337ELDFY/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/JQ64337ELDFY/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/JQ64337ELDFY/image.png 2000w, https://us.v-cdn.net/6038601/uploads/JQ64337ELDFY/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 2. </strong>Case examples for photon-counting detector CT–derived images in adrenal adenoma. Left column: Axial true unenhanced (TUE) images. Middle column: Virtual unenhanced (VUE) images. Right column: Corresponding contrast-enhanced portal venous phase images. (A–C) Images in a 79-year-old woman with lipid-rich adenoma of the left adrenal gland. VUE images led to overestimation of adrenal gland attenuation, which hampered correct classification as lipid-rich adenoma using the 10-HU threshold. (D–F) Images in a 65-year-old woman with lipid-rich adenoma in the left adrenal gland. Classification based on VUE images was in concordance with classification based on TUE images, albeit with a slight overestimation of attenuation on VUE images. (G–I) Images in a 79-year-old woman with lipid-poor adenoma in the left adrenal gland show high contrast material enhancement. VUE images led to 12-HU underestimation when compared with TUE attenuation, which would still result in the correct classification.</p><p></p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Abdomen / adrenal glands</p></td></tr><tr><td><p><strong>Scan mode</strong></p></td><td><p>True unenhanced + portal venous phase</p></td></tr><tr><td><p><strong>Acquisition mode</strong></p></td><td><p>QunatumPlus</p></td></tr><tr><td><p><strong>Tube voltage [kV]</strong></p></td><td><p>120</p></td></tr><tr><td><p><strong>Effective mAs [mAs]</strong></p></td><td><p>Automated exposure control</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>170</p></td></tr><tr><td><p><strong>Pitch</strong></p></td><td><p>0.8</p></td></tr><tr><td><p><strong>Rotation time [s]</strong></p></td><td><p>0.25–0.5</p></td></tr><tr><td><p><strong>Detector slice collimation</strong></p></td><td><p>144 × 0.4 mm</p></td></tr><tr><td><p><strong>Slice thickness</strong></p></td><td><p>2 mm</p></td></tr><tr><td><p><strong>Matrix</strong></p></td><td><p>512 × 512</p></td></tr><tr><td><p><strong>Iterative reconstruction</strong></p></td><td><p>QIR 4</p></td></tr><tr><td><p><strong>Reconstruction</strong></p></td><td><p>Axial</p></td></tr><tr><td><p><strong>Material decomposition</strong></p></td><td><p>VUE reconstructed from portal venous phase</p></td></tr><tr><td><p><strong>Contrast agent</strong></p></td><td><p>Accupaque 300</p></td></tr><tr><td><p><strong>Contrast phase</strong></p></td><td><p>Portal venous</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p>29 patients, 36 adrenal adenomas</p></td></tr><tr><td><p><strong>VUE vs TUE attenuation</strong></p></td><td><p>9.2 ± 8.6 vs 7.3 ± 8.4 HU; <strong>P = .07</strong></p></td></tr><tr><td><p><strong>10-HU classification</strong></p></td><td><p><strong>10/36 adenomas misclassified</strong> using VUE</p></td></tr><tr><td><p><strong>Lipid-rich adenomas</strong></p></td><td><p>23/36</p></td></tr><tr><td><p><strong>Lipid-poor adenomas</strong></p></td><td><p>13/36</p></td></tr><tr><td><p><strong>Tissue dependence</strong></p></td><td><p>VUE underestimated muscle/portal vein and overestimated fat attenuation</p></td></tr><tr><td><p><strong>Clinical implication</strong></p></td><td><p>VUE cannot yet reliably substitute for TUE when applying the conventional 10-HU threshold</p></td></tr></table></div><p><strong>Source:</strong> Lennartz S, Schoenbeck D, Kröger JR, Borggrefe J, Niehoff JH. <em>Photon-counting CT Material Decomposition: Initial Experience in Assessing Adrenal Adenoma.</em> Radiology. 2023;306(1):202–204. doi:10.1148/radiol.220919. </p>]]>
        </description>
    </item>
    <item>
        <title>(2024) Haag | PURE Calcium PCCT may replace true noncontrast calcium scoring</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/503/2024-haag-pure-calcium-pcct-may-replace-true-noncontrast-calcium-scoring</link>
        <pubDate>Fri, 28 Aug 2026 10:53:33 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">503@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In 170 patients, coronary artery calcium scores derived from PCCT PureCalcium virtual noncontrast images showed excellent correlation with true noncontrast (TNC) calcium scoring (ICC 0.98). Median Agatston scores were not significantly different: 4.8 vs 2.7, respectively (p = 0.99).</p><p></p><p>High correlation did not translate into complete interchangeability. PureCalcium correctly classified 125/170 patients (74%) into CAD-RADS plaque-burden categories, with excellent categorical agreement (κ = 0.88), but discrepancies were concentrated around the clinically important P0/P1 boundary. </p><p></p><p>Eliminating the dedicated TNC acquisition could reduce total cardiac CT radiation exposure by approximately 20%, since the TNC scan contributed 19.7% ± 8.8% of total examination dose. PureCalcium therefore shows considerable potential for single-scan CCTA plus calcium assessment, although caution remains necessary in patients with very low calcium burden.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fpubs.rsna.org%2Fdoi%2F10.1148%2Fradiol.230545" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/K3UZYL4QFW6F\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:191100,&quot;width&quot;:1000,&quot;height&quot;:575,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FK3UZYL4QFW6F%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1680,&quot;dateInserted&quot;:&quot;2026-08-28T10:52:37+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FK3UZYL4QFW6F%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/K3UZYL4QFW6F/image.png" alt="image.png" height="575" width="1000" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/K3UZYL4QFW6F/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/K3UZYL4QFW6F/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/K3UZYL4QFW6F/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/K3UZYL4QFW6F/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/K3UZYL4QFW6F/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/K3UZYL4QFW6F/image.png 2000w, https://us.v-cdn.net/6038601/uploads/K3UZYL4QFW6F/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 2</strong>. Representative images of coronary artery calcifications in true noncontrast coronary artery calcium scoring (CACS) (left column), PureCalcium images from coronary CT angiography (CCTA) (middle column), and contrast-enhanced CCTA (right column). All patients underwent CCTA for evaluation of coronary artery disease, with reconstruction in the axial orientation. <strong>(A)</strong> Images in a 47-year-old female patient with no coronary artery calcifications, P0 in the Coronary Artery Disease Reporting and Data System, or CAD-RADS 2.0. <strong>(B)</strong> Images in a 70-year-old male patient with mild coronary artery calcifications, P1 in CAD-RADS 2.0. <strong>(C)</strong> Images in a 72-year-old female patient with moderate coronary artery calcifications, P2 in CAD-RADS 2.0. <strong>(D)</strong> Images in a 65-year-old male patient with severe coronary artery calcifications, P3 in CAD-RADS 2.0. PureCalcium is a customized algorithm for virtual noncontrast image reconstruction derived from photon-counting CCTA specifically for the purpose of CACS. </p><p></p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha, VA50</p></td></tr><tr><td><p><strong>Software</strong></p></td><td><p>syngo.via VB60</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Coronary arteries</p></td></tr><tr><td><p><strong>TNC scan mode</strong></p></td><td><p>High-pitch</p></td></tr><tr><td><p><strong>CCTA scan mode</strong></p></td><td><p>Flash, sequential, or spiral depending on heart rate/rhythm</p></td></tr><tr><td><p><strong>Acquisition mode</strong></p></td><td><p>Standard-resolution PCCT</p></td></tr><tr><td><p><strong>Tube voltage [kV]</strong></p></td><td><p>120</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>100</p></td></tr><tr><td><p><strong>Dose modulation</strong></p></td><td><p>Automatic tube current modulation</p></td></tr><tr><td><p><strong>Matrix</strong></p></td><td><p>512 × 512</p></td></tr><tr><td><p><strong>TNC slice thickness / increment [mm]</strong></p></td><td><p>3.0 / 1.5</p></td></tr><tr><td><p><strong>CCTA slice thickness / increment [mm]</strong></p></td><td><p>0.4 / 0.2</p></td></tr><tr><td><p><strong>TNC kernel</strong></p></td><td><p>Qr36</p></td></tr><tr><td><p><strong>TNC iterative reconstruction</strong></p></td><td><p>None</p></td></tr><tr><td><p><strong>TNC reconstruction energy</strong></p></td><td><p>70 keV</p></td></tr><tr><td><p><strong>PureCalcium slice thickness / increment [mm]</strong></p></td><td><p>2.0 / 1.5</p></td></tr><tr><td><p><strong>PureCalcium kernel</strong></p></td><td><p>Qr40</p></td></tr><tr><td><p><strong>PureCalcium iterative reconstruction</strong></p></td><td><p>QIR 3</p></td></tr><tr><td><p><strong>Calcium scoring thickness</strong></p></td><td><p>Automatic resampling to 1.5 mm</p></td></tr><tr><td><p><strong>Calcium threshold</strong></p></td><td><p>130 HU; minimum area 0.5 mm²</p></td></tr><tr><td><p><strong>Contrast agent</strong></p></td><td><p>Iohexol, Accupaque 300</p></td></tr><tr><td><p><strong>Contrast volume</strong></p></td><td><p>50–65 mL</p></td></tr><tr><td><p><strong>Injection rate</strong></p></td><td><p>4 mL/s</p></td></tr><tr><td><p><strong>Contrast timing</strong></p></td><td><p>Test bolus</p></td></tr><tr><td><p><strong>Heart-rate control</strong></p></td><td><p>Up to 15 mg IV metoprolol if required</p></td></tr><tr><td><p><strong>Vasodilation</strong></p></td><td><p>0.4 mg sublingual nitroglycerin if not contraindicated</p></td></tr></table></div><p></p><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p><strong>170 patients</strong>; 111 with CACS &gt;0</p></td></tr><tr><td><p><strong>Median Agatston score</strong></p></td><td><p>TNC <strong>4.8</strong> vs PureCalcium <strong>2.7</strong>, <em>p</em> = 0.99</p></td></tr><tr><td><p><strong>Score correlation</strong></p></td><td><p><strong>ICC 0.98</strong> (95% CI 0.97–0.99)</p></td></tr><tr><td><p><strong>Plaque-burden classification</strong></p></td><td><p><strong>125/170 (74%)</strong> correctly classified; κ = <strong>0.88</strong></p></td></tr><tr><td><p><strong>P0/P1 discrepancies</strong></p></td><td><p>20 patients classified P1 instead of P0; 14 classified P0 instead of P1</p></td></tr><tr><td><p><strong>Radiation contribution of TNC</strong></p></td><td><p>DLP <strong>32.5 ± 11.5 mGy·cm</strong>; <strong>19.7% ± 8.8%</strong> of total examination dose</p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Haag NP, Michael AE, Lennartz S, Panknin C, Niehoff JH, Borggrefe J, Shahzadi I, Zwanenburg A, Kroeger JR. <em>Coronary Artery Calcium Scoring Using Virtual Versus True Noncontrast Images From Photon-Counting Coronary CT Angiography.</em> <strong>Radiology.</strong> 2024;310(3). doi:10.1148/radiol.230545.</p>]]>
        </description>
    </item>
    <item>
        <title>(2024) Gnasso | Optimized PCCT reconstruction improves myocardial ECV accuracy</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/502/2024-gnasso-optimized-pcct-reconstruction-improves-myocardial-ecv-accuracy</link>
        <pubDate>Fri, 28 Aug 2026 10:48:54 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">502@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In 27 patients undergoing same-day PCCT and cardiac MRI, myocardial extracellular volume (ECV) measurements remained robust across a wide range of reconstruction settings, but reconstruction choices meaningfully affected agreement with MRI.</p><p></p><p>The greatest improvement came from combining thin slices with low-keV VMI, rather than from changing iterative reconstruction strength alone. Slice thickness was a particularly important determinant of ECV accuracy. Reconstruction at 0.4/0.3 mm thickness/increment achieved the highest agreement with MRI (CCC ≈ 0.94) and substantially reduced bias compared with thicker reconstructions, likely by limiting partial-volume averaging within the myocardium.</p><p></p><p>The optimal combination was QIR 4, 0.4-mm slices with 0.3-mm increment, and 45-keV VMI. Compared with the literature-standard reconstruction of QIR 3, 1.5 mm, and 65 keV, this reduced mean ECV bias by 63% and increased concordance with MRI by 6%, bringing PCCT-derived ECV substantially closer to the MRI reference.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Flink.springer.com%2Farticle%2F10.1186%2Fs41747-024-00469-7" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/KS1C1846OL3N\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:1246377,&quot;width&quot;:1944,&quot;height&quot;:1036,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FKS1C1846OL3N%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1679,&quot;dateInserted&quot;:&quot;2026-08-28T10:47:38+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FKS1C1846OL3N%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/KS1C1846OL3N/image.png" alt="image.png" height="1036" width="1944" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/KS1C1846OL3N/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/KS1C1846OL3N/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/KS1C1846OL3N/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/KS1C1846OL3N/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/KS1C1846OL3N/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/KS1C1846OL3N/image.png 2000w, https://us.v-cdn.net/6038601/uploads/KS1C1846OL3N/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 1. </strong>Representation of the systematic step-by-step protocol optimization approach. In Step 1, the optimal quantum iterative reconstruction level (14) was identified. In Step 2, the different slice thicknesses were tested from 0.4 to 8 mm. In Step 3, the virtual monoenergetic image reconstruction level of the late iodine enhancement scan was optimized from 40 to 90 keV. The parameter with the best diagnostic performance in each step was retained in the subsequent step.</p><p></p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Cardiac / myocardium</p></td></tr><tr><td><p><strong>Scan mode</strong></p></td><td><p>QuantumPlus</p></td></tr><tr><td><p><strong>Acquisition mode</strong></p></td><td><p>PCCT spectral cardiac acquisition</p></td></tr><tr><td><p><strong>Tube voltage [kV]</strong></p></td><td><p>120</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>44 CCTA; 55 late iodine enhancement</p></td></tr><tr><td><p><strong>Dose modulation</strong></p></td><td><p>CARE Dose4D</p></td></tr><tr><td><p><strong>Rotation time [s]</strong></p></td><td><p>0.25</p></td></tr><tr><td><p><strong>Detector slice collimation</strong></p></td><td><p>144 × 0.4 mm</p></td></tr><tr><td><p><strong>ECG gating</strong></p></td><td><p>Prospective ECG-triggered</p></td></tr><tr><td><p><strong>CCTA ECG window</strong></p></td><td><p>30–80% of RR interval</p></td></tr><tr><td><p><strong>Late iodine enhancement timing</strong></p></td><td><p>5 min after contrast administration</p></td></tr><tr><td><p><strong>Late iodine enhancement ECG timing</strong></p></td><td><p>Fixed 280-ms delay from R-wave</p></td></tr><tr><td><p><strong>Baseline slice thickness / increment [mm]</strong></p></td><td><p>1.5 / 1.0</p></td></tr><tr><td><p><strong>Baseline kernel</strong></p></td><td><p>Qr40</p></td></tr><tr><td><p><strong>Baseline iterative reconstruction</strong></p></td><td><p>QIR 3</p></td></tr><tr><td><p><strong>Baseline reconstruction energy</strong></p></td><td><p>65 keV</p></td></tr><tr><td><p><strong>QIR levels tested</strong></p></td><td><p>QIR 1–4</p></td></tr><tr><td><p><strong>Slice thickness / increment tested [mm]</strong></p></td><td><p>0.4/0.3; 1.5/1.0; 3.0/2.5; 5.0/4.0; 8.0/7.0</p></td></tr><tr><td><p><strong>VMI energies tested</strong></p></td><td><p>40, 45, 50, 55, 60, 65, 70, 90 keV</p></td></tr><tr><td><p><strong>Optimal reconstruction</strong></p></td><td><p><strong>QIR 4; 0.4/0.3 mm; 45 keV</strong></p></td></tr><tr><td><p><strong>ECV reconstruction</strong></p></td><td><p>Iodine maps generated from late iodine enhancement scan</p></td></tr><tr><td><p><strong>Contrast agent</strong></p></td><td><p>Iopromide, Ultravist 370 mg I/mL</p></td></tr><tr><td><p><strong>Contrast volume</strong></p></td><td><p>100 mL</p></td></tr><tr><td><p><strong>Saline flush</strong></p></td><td><p>20 mL</p></td></tr><tr><td><p><strong>Injection rate</strong></p></td><td><p>5–6 mL/s</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p>27 patients undergoing same-day PCCT and MRI</p></td></tr><tr><td><p><strong>QIR effect</strong></p></td><td><p>No significant ECV difference across QIR 1–4, <em>p</em> = 1.000</p></td></tr><tr><td><p><strong>Best slice thickness</strong></p></td><td><p><strong>0.4 mm</strong>, CCC ≈ <strong>0.94</strong>, bias <strong>0.8%</strong></p></td></tr><tr><td><p><strong>Best VMI energy</strong></p></td><td><p><strong>45 keV</strong>, mean bias <strong>0.6%</strong></p></td></tr><tr><td><p><strong>Optimal reconstruction</strong></p></td><td><p><strong>QIR 4 + 0.4/0.3 mm + 45 keV</strong></p></td></tr><tr><td><p><strong>Improvement vs baseline</strong></p></td><td><p><strong>63% lower mean bias</strong> and <strong>6% higher concordance</strong> with MRI</p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Gnasso C, Pinos D, Schoepf UJ, Vecsey-Nagy M, Aquino GJ, Fink N, Zsarnoczay E, Holtackers RJ, Stock J, Suranyi P, Varga-Szemes A, Emrich T. <em>Impact of reconstruction parameters on the accuracy of myocardial extracellular volume quantification on a first-generation, photon-counting detector CT.</em> <strong>European Radiology Experimental.</strong> 2024;8:70. doi:10.1186/s41747-024-00469-7.</p>]]>
        </description>
    </item>
    <item>
        <title>(2024) Fink | Safety-net PCCT VNC improves subtle coronary calcium detection</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/501/2024-fink-safety-net-pcct-vnc-improves-subtle-coronary-calcium-detection</link>
        <pubDate>Fri, 28 Aug 2026 10:44:12 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">501@/de/discussions</guid>
        <description><![CDATA[<div><div><p>Standard PCCT virtual noncontrast (VNC) calcium scoring performs well for larger and denser plaques but can miss subtle disease. Large/medium and high/medium-density plaques showed strong agreement with true noncontrast (TNC) scoring (r &gt; 0.9; ICC ≥ 0.90), whereas low-density plaques showed poor agreement and small plaques were not detectable with standard VNC reconstruction.</p><p></p><p>A dedicated “safety-net” reconstruction combining thinner 1.0-mm slices, a sharper Qr44 kernel, low monoenergetic energy and lower calcium thresholds markedly improved subtle-plaque detection. The best clinical setting was 55 keV, QIR 2, and a 110-HU threshold.</p><p></p><p>In 63 participants, standard VNC missed 56 plaques visible on TNC; the optimal safety-net reconstruction recovered 50/56 (89%). The approach may therefore be particularly valuable when standard VNC yields a calcium score of zero, potentially preventing false-negative risk classification while supporting elimination of a separate TNC scan.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fpubs.rsna.org%2Fdoi%2F10.1148%2Fryct.230328" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/WLH9CD6E09FF\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:1919996,&quot;width&quot;:2075,&quot;height&quot;:1338,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FWLH9CD6E09FF%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1677,&quot;dateInserted&quot;:&quot;2026-08-28T10:40:55+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FWLH9CD6E09FF%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/WLH9CD6E09FF/image.png" alt="image.png" height="1338" width="2075" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/WLH9CD6E09FF/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/WLH9CD6E09FF/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/WLH9CD6E09FF/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/WLH9CD6E09FF/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/WLH9CD6E09FF/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/WLH9CD6E09FF/image.png 2000w, https://us.v-cdn.net/6038601/uploads/WLH9CD6E09FF/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 7</strong>: Example axial CT reconstructions of in vivo coronary artery calcium scoring (CACS) without and with calcium scoring overlay, comparing CACSTNC (left column: 70 keV, quantum iterative reconstruction [QIR] off) and CACSVNC (middle column: 55 keV, QIR 1, as previously recommended) (9) at standard CACS settings with CACSVNC at the proposed safety net reconstruction (right column: 55 keV, QIR 2). The upper row (participant 1) shows CT scans in a 90-year-old male, the middle row (participant 2) in a 62-year-old male, and the lower row (participant 3) in a 59-year-old female. The safety net reconstruction allowed improved detection of small and/or low-density in vivo plaques (arrowheads) that initially were detected in reference TNC scans but not in standard VNC reconstructions. LAD = left anterior descending artery, RCA = right coronary artery, TNC = true noncontrast, VNC = virtual noncontrast. </p><p></p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha, VA40</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Coronary arteries / cardiac CT</p></td></tr><tr><td><p><strong>Scan mode</strong></p></td><td><p>QuantumPlus</p></td></tr><tr><td><p><strong>Tube voltage [kV]</strong></p></td><td><p>120</p></td></tr><tr><td><p><strong>Rotation time [s]</strong></p></td><td><p>0.25</p></td></tr><tr><td><p><strong>Detector slice collimation</strong></p></td><td><p>144 × 0.4 mm</p></td></tr><tr><td><p><strong>ECG gating</strong></p></td><td><p>ECG-triggered</p></td></tr><tr><td><p><strong>Reference TNC reconstruction</strong></p></td><td><p>70 keV, QIR off</p></td></tr><tr><td><p><strong>Standard TNC slice thickness / increment [mm]</strong></p></td><td><p>3.0 / 1.5</p></td></tr><tr><td><p><strong>Standard TNC kernel</strong></p></td><td><p>Qr36</p></td></tr><tr><td><p><strong>VNC reconstruction energies</strong></p></td><td><p>55–80 keV in 5-keV increments</p></td></tr><tr><td><p><strong>VNC iterative reconstruction</strong></p></td><td><p>QIR 1–4</p></td></tr><tr><td><p><strong>Standard VNC slice thickness</strong></p></td><td><p>3.0 mm</p></td></tr><tr><td><p><strong>Standard VNC kernel</strong></p></td><td><p>Qr36</p></td></tr><tr><td><p><strong>Safety-net slice thickness / increment [mm]</strong></p></td><td><p>1.0 / 1.0</p></td></tr><tr><td><p><strong>Safety-net kernel</strong></p></td><td><p>Qr44</p></td></tr><tr><td><p><strong>Preferred safety-net reconstruction</strong></p></td><td><p><strong>55 keV, QIR 2</strong></p></td></tr><tr><td><p><strong>Preferred calcium threshold</strong></p></td><td><p><strong>110 HU</strong></p></td></tr><tr><td><p><strong>Contrast agent</strong></p></td><td><p>Iopromide 350 mg I/mL</p></td></tr><tr><td><p><strong>Contrast protocol</strong></p></td><td><p>50 mL contrast → 20 mL 50% contrast/saline → 25 mL saline</p></td></tr><tr><td><p><strong>Bolus triggering</strong></p></td><td><p>Descending aorta, 150 HU</p></td></tr><tr><td><p><strong>Trigger delay</strong></p></td><td><p>8 s</p></td></tr><tr><td><p><strong>Patient preparation</strong></p></td><td><p>0.4 mg sublingual nitroglycerin; 5 mg oral metoprolol if HR &gt;70 bpm</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Clinical cohort</strong></p></td><td><p>63 participants; mean age <strong>57.8 ± 15.5 years</strong></p></td></tr><tr><td><p><strong>Large/medium or denser plaques</strong></p></td><td><p>VNC vs TNC: <strong>r &gt; 0.9; ICC ≥ 0.90</strong></p></td></tr><tr><td><p><strong>Low-density plaques, standard VNC</strong></p></td><td><p>Poor agreement: <strong>r = 0.5–0.6; ICC = 0.20–0.48</strong></p></td></tr><tr><td><p><strong>Small plaques, standard VNC</strong></p></td><td><p><strong>Not detectable</strong></p></td></tr><tr><td><p><strong>Best safety-net reconstruction</strong></p></td><td><p><strong>55 keV, QIR 2, Qr44, 1.0 mm, 110 HU</strong></p></td></tr><tr><td><p><strong>Clinical plaque recovery</strong></p></td><td><p><strong>50/56 (89%)</strong> previously missed plaques detected</p></td></tr></table></div><p></p><p><strong>Source:</strong><br />
Fink N, Emrich T, Schoepf UJ, Zsarnoczay E, O’Doherty J, Halfmann MC, Griffith JP III, et al. <em>Improved Detection of Small and Low-Density Plaques in Virtual Noncontrast Imaging–based Calcium Scoring at Photon-Counting Detector CT.</em> <strong>Radiology: Cardiothoracic Imaging.</strong> 2024;6(4). doi:10.1148/ryct.230328.</p>]]>
        </description>
    </item>
    <item>
        <title>(2024) Augustin | Sharp-kernel PCCT improves lower-extremity CTA vs DSA</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/499/2024-augustin-sharp-kernel-pcct-improves-lower-extremity-cta-vs-dsa</link>
        <pubDate>Fri, 28 Aug 2026 10:31:44 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">499@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In 39 patients with 45 lower-extremity run-off studies, PCCT angiography was compared directly with digital subtraction angiography (DSA) for peripheral arterial occlusive disease. The cohort was challenging, with 77.8% of studies representing Fontaine stage IV disease, making severe calcification and small crural vessels particularly relevant. </p><p></p><p>Reconstruction kernel selection had a major effect on diagnostic performance. Increasing sharpness from Bv36 → Bv48 → Bv56 progressively reduced calcium blooming and improved vessel sharpness and agreement with DSA, while sensitivity for ≥60% stenosis remained essentially unchanged at approximately 81%. </p><p></p><p>The benefit was strongest in severely calcified crural arteries: specificity improved from 49.3% with Bv36 to 65.9% with Bv56, while stenosis overestimation decreased and diagnostic confidence increased. Thus, for advanced lower-extremity PAOD, the sharper Bv56 kernel provided the best diagnostic trade-off despite increased image noise.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fwww.academicradiology.org%2Farticle%2FS1076-6332%2824%2900078-3%2Ffulltext" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/9DCXW48X6P9M\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:229611,&quot;width&quot;:1000,&quot;height&quot;:575,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F9DCXW48X6P9M%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1676,&quot;dateInserted&quot;:&quot;2026-08-28T10:29:23+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F9DCXW48X6P9M%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/9DCXW48X6P9M/image.png" alt="image.png" height="575" width="1000" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/9DCXW48X6P9M/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/9DCXW48X6P9M/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/9DCXW48X6P9M/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/9DCXW48X6P9M/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/9DCXW48X6P9M/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/9DCXW48X6P9M/image.png 2000w, https://us.v-cdn.net/6038601/uploads/9DCXW48X6P9M/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 6. </strong>Imaging of the crural vessels in a 95-year-old woman with PAOD stage 4 and necrosis of both thighs. Axial PCD-CTA images of the crural vessels at the level of the distal lower thigh reconstructed with different spatial frequencies on the left side. With Bv36, despite low image noise, the lack of vessel sharpness hampers contrast assessment in the anterior tibial artery. The sharper Bv48 and Bv56 kernels reveal significant wall calcifications and an atheromatous plaque. Meanwhile, reduction of calcium blooming allows for detailed luminal assessment. Corresponding DSA image of the distal run-off confirms a short high-grade stenosis of the distal anterior tibial artery (arrow), while the posterior tibial artery is completely occluded.</p><p></p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Abdominal aorta and lower-extremity run-off</p></td></tr><tr><td><p><strong>Scan mode</strong></p></td><td><p>Standard resolution, QuantumPlus</p></td></tr><tr><td><p><strong>Acquisition mode</strong></p></td><td><p>PCCT</p></td></tr><tr><td><p><strong>Tube voltage [kV]</strong></p></td><td><p>140</p></td></tr><tr><td><p><strong>Effective mAs [mAs]</strong></p></td><td><p>CARE Dose4D automatic</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>26</p></td></tr><tr><td><p><strong>Dose modulation</strong></p></td><td><p>CARE Dose4D</p></td></tr><tr><td><p><strong>CTDIvol [mGy]</strong></p></td><td><p>6.46 ± 0.80</p></td></tr><tr><td><p><strong>DLP [mGy·cm]</strong></p></td><td><p>857.77 ± 120.17</p></td></tr><tr><td><p><strong>Rotation time [s]</strong></p></td><td><p>0.5</p></td></tr><tr><td><p><strong>Pitch</strong></p></td><td><p>0.4</p></td></tr><tr><td><p><strong>Matrix</strong></p></td><td><p>512 × 512</p></td></tr><tr><td><p><strong>Field of view [mm]</strong></p></td><td><p>150 × 150 per leg</p></td></tr><tr><td><p><strong>Detector slice collimation</strong></p></td><td><p>144 × 0.4 mm</p></td></tr><tr><td><p><strong>Slice thickness [mm]</strong></p></td><td><p>0.8</p></td></tr><tr><td><p><strong>Reconstruction increment [mm]</strong></p></td><td><p>0.8</p></td></tr><tr><td><p><strong>Reconstruction kernel</strong></p></td><td><p>Bv36, Bv48, Bv56</p></td></tr><tr><td><p><strong>Iterative reconstruction</strong></p></td><td><p>QIR 3</p></td></tr><tr><td><p><strong>Reconstruction</strong></p></td><td><p>Axial, each leg reconstructed separately</p></td></tr><tr><td><p><strong>Contrast agent</strong></p></td><td><p>Imeron 350</p></td></tr><tr><td><p><strong>Contrast volume</strong></p></td><td><p>90 mL</p></td></tr><tr><td><p><strong>Saline flush</strong></p></td><td><p>20 mL pre-bolus + 20 mL post-bolus</p></td></tr><tr><td><p><strong>Injection rate</strong></p></td><td><p>3.0 mL/s</p></td></tr><tr><td><p><strong>Bolus triggering</strong></p></td><td><p>Abdominal aorta, 140-HU threshold</p></td></tr><tr><td><p><strong>Trigger delay</strong></p></td><td><p>18 s</p></td></tr></table></div><p></p><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Cohort</strong></p></td><td><p>39 patients / 45 PCCT run-off studies; <strong>77.8% Fontaine IV</strong></p></td></tr><tr><td><p><strong>Sensitivity for ≥60% stenosis</strong></p></td><td><p>Bv36 <strong>81.6%</strong>; Bv48 <strong>81.5%</strong>; Bv56 <strong>81.0%</strong></p></td></tr><tr><td><p><strong>Specificity</strong></p></td><td><p>Increased from <strong>71.1% → 76.9% → 79.6%</strong> with Bv36 → Bv48 → Bv56</p></td></tr><tr><td><p><strong>Severely calcified crural vessels</strong></p></td><td><p>Specificity improved from <strong>49.3% with Bv36 to 65.9% with Bv56</strong>, <em>p</em> = 0.024</p></td></tr><tr><td><p><strong>Agreement with DSA</strong></p></td><td><p>Bv56 significantly improved stenosis concordance vs Bv36, <strong>p = 0.002</strong></p></td></tr><tr><td><p><strong>Diagnostic confidence</strong></p></td><td><p>Highest with <strong>Bv56</strong>, reflecting reduced calcium blooming and improved vessel sharpness</p></td></tr></table></div><p><strong>Source:</strong><br />
Augustin AM, Hartung V, Grunz JP, et al. <em>Photon-Counting Detector CT Angiography Versus Digital Subtraction Angiography in Patients with Peripheral Arterial Disease.</em> <strong>Academic Radiology.</strong> 2024;31(7):2973–2986. doi:10.1016/j.acra.2024.02.008.</p>]]>
        </description>
    </item>
    <item>
        <title>(2023) Aquino | PCCT ECV mapping correlates strongly with cardiac MRI</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/498/2023-aquino-pcct-ecv-mapping-correlates-strongly-with-cardiac-mri</link>
        <pubDate>Fri, 28 Aug 2026 10:27:35 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">498@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In this prospective same-day comparison of 29 patients, myocardial extracellular volume (ECV) measured with PCCT showed strong agreement with cardiac MRI, supporting spectral PCCT as a quantitative tool for myocardial tissue characterization. </p><p></p><p>Dual-energy PCCT achieved correlations of r = 0.91 for global ECV and r = 0.82 for midventricular ECV versus MRI. Dual-energy PCCT slightly overestimated ECV by approximately 2%, whereas the single-energy approach underestimated ECV by approximately 3%. Despite this small systematic bias, reliability versus MRI was good to excellent, with ICCs of 0.90 globally and 0.81 at the midventricular level for dual-energy PCCT. </p><p></p><p>A major advantage of the spectral approach was that dual-energy ECV required only the late-enhancement acquisition, eliminating the true unenhanced scan needed for single-energy ECV. This reduced radiation exposure by 40% (CTDIvol 10.1 vs 16.8 mGy) while also reducing potential interscan registration errors.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fpubs.rsna.org%2Fdoi%2F10.1148%2Fradiol.222030" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/WZ2TEENIR7XI\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:639189,&quot;width&quot;:1000,&quot;height&quot;:575,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FWZ2TEENIR7XI%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1675,&quot;dateInserted&quot;:&quot;2026-08-28T10:26:07+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FWZ2TEENIR7XI%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/WZ2TEENIR7XI/image.png" alt="image.png" height="575" width="1000" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/WZ2TEENIR7XI/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/WZ2TEENIR7XI/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/WZ2TEENIR7XI/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/WZ2TEENIR7XI/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/WZ2TEENIR7XI/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/WZ2TEENIR7XI/image.png 2000w, https://us.v-cdn.net/6038601/uploads/WZ2TEENIR7XI/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Figure 3.</strong> Images show subepicardial late enhancement in a 50-year-old woman with history of arrhythmogenic right ventricular cardiomyopathy. Short- and long-axis cardiac MRI and CT scans demonstrate patchy left ventricular midmyocardial and subepicardial involvement. Enhancement pattern may suggest arrhythmogenic right ventricular cardiomyopathy with left ventricular involvement, although remote myocarditis cannot be ruled out. (A–C) Top row: Images show (A) basal and (B) midventricular short-axis late gadolinium enhancement (LGE) MRI scans with subepicardial involvement (arrows), similarly in (C) the long-axis image. (D–F) Middle row: Images are corresponding CT extracellular volume (ECV) maps showing a similar pattern of myocardial involvement (arrows). (G–I) Bottom row: Images show (G, H) corresponding, short-axis MRI ECV maps demonstrating similar areas of involvement (arrows), and a (I) representative CT ECV American Heart Association 17-segment subepicardiac (50% to outer 10% of the myocardial wall) polar map, which shows the hematocrit (Hct) and the global ECV mean, as well as the means and SDs of ECV for each myocardial segment.</p><p></p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Cardiac / myocardium</p></td></tr><tr><td><p><strong>Scan mode</strong></p></td><td><p>QuantumPlus sequential</p></td></tr><tr><td><p><strong>Acquisition mode</strong></p></td><td><p>PCCT, full spectral data</p></td></tr><tr><td><p><strong>Scan direction</strong></p></td><td><p>Axial</p></td></tr><tr><td><p><strong>Tube voltage [kV]</strong></p></td><td><p>120</p></td></tr><tr><td><p><strong>Effective mAs [mAs]</strong></p></td><td><p>CARE Dose4D automatic</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>20 unenhanced; 44 CCTA; 50 late enhancement</p></td></tr><tr><td><p><strong>Dose modulation</strong></p></td><td><p>CARE Dose4D</p></td></tr><tr><td><p><strong>CTDIvol [mGy]</strong></p></td><td><p>Dual-energy ECV: 10.1; single-energy ECV: 16.8</p></td></tr><tr><td><p><strong>DLP [mGy·cm]</strong></p></td><td><p>Dual-energy ECV: 148; single-energy ECV: 248</p></td></tr><tr><td><p><strong>Rotation time [s]</strong></p></td><td><p>0.25</p></td></tr><tr><td><p><strong>Acquisition slice thickness [mm]</strong></p></td><td><p>0.4</p></td></tr><tr><td><p><strong>Reconstruction slice width [mm]</strong></p></td><td><p>1.5</p></td></tr><tr><td><p><strong>Reconstruction increment [mm]</strong></p></td><td><p>1.0</p></td></tr><tr><td><p><strong>Reconstruction kernel</strong></p></td><td><p>Qr40</p></td></tr><tr><td><p><strong>Iterative reconstruction</strong></p></td><td><p>QIR 3</p></td></tr><tr><td><p><strong>Reconstruction energy</strong></p></td><td><p>65 keV</p></td></tr><tr><td><p><strong>Reconstruction comparison</strong></p></td><td><p>Single-energy ECV vs dual-energy iodine-based ECV</p></td></tr><tr><td><p><strong>ECG gating</strong></p></td><td><p>ECG-triggered</p></td></tr><tr><td><p><strong>ECG window</strong></p></td><td><p>280 ms RR interval for unenhanced and late enhancement; 30–80% RR for CCTA</p></td></tr><tr><td><p><strong>Contrast agent</strong></p></td><td><p>Iopromide, Ultravist 370 mg I/mL</p></td></tr><tr><td><p><strong>Contrast volume</strong></p></td><td><p>100 mL</p></td></tr><tr><td><p><strong>Saline flush</strong></p></td><td><p>20 mL</p></td></tr><tr><td><p><strong>Injection rate</strong></p></td><td><p>5–6 mL/s</p></td></tr><tr><td><p><strong>Bolus triggering</strong></p></td><td><p>n.a.</p></td></tr><tr><td><p><strong>Late-enhancement phase</strong></p></td><td><p>5 min after contrast administration</p></td></tr><tr><td><p><strong>ECV reconstruction</strong></p></td><td><p>Dual-energy iodine maps from late-enhancement scan; automated 17-segment ECV maps</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p>29 patients</p></td></tr><tr><td><p><strong>Global ECV, dual-energy PCCT vs MRI</strong></p></td><td><p><strong>r = 0.91; ICC = 0.90</strong></p></td></tr><tr><td><p><strong>Midventricular ECV</strong></p></td><td><p><strong>r = 0.82; ICC = 0.81</strong></p></td></tr><tr><td><p><strong>Dual-energy bias</strong></p></td><td><p>ECV overestimated by <strong>1.9% globally</strong></p></td></tr><tr><td><p><strong>Radiation reduction</strong></p></td><td><p><strong>40%</strong> with dual- vs single-energy ECV</p></td></tr><tr><td><p><strong>Clinical implication</strong></p></td><td><p>Spectral PCCT may add quantitative myocardial tissue characterization to cardiac CT without requiring an additional unenhanced scan</p></td></tr></table></div><p><strong>Source:</strong><br />
Aquino GJ, O’Doherty J, Schoepf UJ, Ellison B, Byrne J, Fink N, Zsarnoczay E, et al. <em>Myocardial Characterization with Extracellular Volume Mapping with a First-Generation Photon-counting Detector CT with MRI Reference.</em> <strong>Radiology.</strong> 2023;307(2). doi:10.1148/radiol.222030.</p>]]>
        </description>
    </item>
    <item>
        <title>(2023) Schade | Pseudoenhancement of renal cysts on PCCT</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/495/2023-schade-pseudoenhancement-of-renal-cysts-on-pcct</link>
        <pubDate>Fri, 28 Aug 2026 10:10:05 +0000</pubDate>
        <category>Scientific publications</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">495@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In 86 patients with 160 non-enhancing renal cysts, the study assessed how PCCT virtual monoenergetic image (VMI) energy level affects apparent renal cyst enhancement and therefore lesion classification.</p><p><br />
Low-keV VMI substantially increased apparent cyst attenuation: mean attenuation rose from 4 ± 11 HU at 190 keV to 36 ± 17 HU at 40 keV.</p><p><br />
This produced marked pseudoenhancement at low energies: at 40 keV, 35% of cysts appeared equivocally enhancing and 46% definitely enhancing; at 50 keV, these rates were 29% and 10%, respectively. At 60 keV, 9% were equivocal but none definitely enhancing.</p><p><br />
Importantly, no cyst was classified as enhancing at 70–190 keV, indicating that ≥70-keV VMI provides substantially more reliable assessment of true renal cyst enhancement.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fwww.academicradiology.org%2Farticle%2FS1076-6332%2823%2900187-3%2Fabstract" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><p><strong>Practical PCCT takeaway: </strong></p><p>Low-keV reconstructions are attractive for increasing iodine conspicuity, but they should not be used uncritically for HU-based renal lesion characterization. For renal cyst enhancement, 70 keV appears to be a safer reference reconstruction in this study.</p><p>The observed misclassification was not significantly dependent on cyst size, location, or patient BMI, suggesting that the VMI energy itself is the dominant issue.</p><p><strong>Protocol:</strong> </p><p>Clinical PCCT, 120 kVp, image-quality level 68; triphasic examination. Venous-phase VMIs were reconstructed from 40–190 keV in 10-keV increments. Only cysts ≥7 mm confirmed as non-enhancing (&lt;10 HU difference between unenhanced and venous 70-keV images) were analyzed.</p><p>Source: Schade KA, Mergen V, Sartoretti T, Alkadhi H, Euler A. Pseudoenhancement in Cystic Renal Lesions – Impact of Virtual Monoenergetic Images of Photon-Counting Detector CT on Lesion Classification. Academic Radiology. 2023;30(Suppl 1):S305-S313. doi:10.1016/j.acra.2023.04.005. Published online May 5, 2023.</p>]]>
        </description>
    </item>
    <item>
        <title>(2025) Dang | Low-dose 70 kV vs 120kV PCCT for infant cardiac imaging</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/493/2025-dang-low-dose-70-kv-vs-120kv-pcct-for-infant-cardiac-imaging</link>
        <pubDate>Fri, 28 Aug 2026 09:42:55 +0000</pubDate>
        <category>Scientific publications</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">493@/de/discussions</guid>
        <description><![CDATA[<div><div><p>In 58 infants &lt;10 kg, cardiac PCCT at 70 kV and 120 kV provided comparable image quality, with similar aortic CNR and median subjective image-quality scores of 5/5 in both groups.</p><p></p><p>70 kV reduced radiation dose by almost 50%, with CTDI decreasing from 0.25 mGy at 120 kV to 0.13 mGy at 70 kV (P&lt;0.001).</p><p></p><p>The key trade-off is that 70 kV does not provide the advanced spectral reconstructions available at 120 kV, making protocol selection a balance between minimum dose and spectral capability.</p></div></div><p> <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fdoi.org%2F10.1007%2Fs00247-025-06467-0" rel="nofollow noopener ugc" role="button">Access publication</a></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/6D7EHL5COHGN\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:271281,&quot;width&quot;:685,&quot;height&quot;:390,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F6D7EHL5COHGN%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1672,&quot;dateInserted&quot;:&quot;2026-08-28T09:39:40+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F6D7EHL5COHGN%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/6D7EHL5COHGN/image.png" alt="image.png" height="390" width="685" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/6D7EHL5COHGN/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/6D7EHL5COHGN/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/6D7EHL5COHGN/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/6D7EHL5COHGN/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/6D7EHL5COHGN/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/6D7EHL5COHGN/image.png 2000w, https://us.v-cdn.net/6038601/uploads/6D7EHL5COHGN/image.png" sizes="100vw" /></a>
    </span>
</span>
<p><strong>Source:</strong> Dang NH, Zhou W, Arguello Fletes GM, <em>et al.</em> <em>Pediatr Radiol.</em> Published online November 18, 2025. doi:10.1007/s00247-025-06467-0.</p><p></p><p>Furthermore, there is a commentary by Kelly Horst and Lifeng Yu.</p><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Flink.springer.com%2Farticle%2F10.1007%2Fs00247-025-06479-w" rel="nofollow noopener ugc" role="button">Access publication</a></p><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Flink.springer.com%2Fepdf%2F10.1007%2Fs00247-025-06479-w%3Fsharing_token%3Dqw-J6-5uxrkwKTPMSFTFmve4RwlQNchNByi7wbcMAY7-1I44Z-zYplFNhiC2hhR9i1yg0ZNFCaNSoM4WdMvbHUbW0F9Jkst-gl4qGFJQ6ppnbYu_OteyA6lV8QMY2qpX2JNsxrZ6Iq61Z23kvZbpXtZesdKNHSphB32VHFkFu2Q%253D" rel="nofollow noopener ugc" role="button">PDF commentary</a></p><div><div><p>With modern <strong>PCCT</strong>, pediatric cardiac CT can already achieve <strong>sub-mSv radiation doses</strong>; the commentary argues that further dose reduction should not automatically take priority when it compromises <strong>spatial resolution, image noise, edge sharpness, or diagnostic confidence</strong>.</p><p></p><p>Lower-kV PCCT protocols may reduce radiation dose, but there are important trade-offs: <strong>70–90 kV Quantum mode can sacrifice ultra-high spatial resolution and material-specific spectral imaging</strong>, whereas <strong>120–140 kV Quantum Plus mode</strong> enables low-keV VMI and spectral reconstructions such as virtual non-calcium and iodine-only imaging—potentially valuable for cardiovascular assessment.</p><p></p><p>The authors propose a shift from <strong>ALARA toward optimizing diagnostic performance and clinical impact</strong>: PCCT protocols should preserve sufficient image quality not only for human interpretation, but also for <strong>quantitative imaging, radiomics and AI</strong>, whose performance and reproducibility may deteriorate as radiation dose decreases and image noise increases.</p></div></div><p><strong>Source:</strong> Horst K, Yu L. <em>Pediatr Radiol.</em> 2025;56:554–555.</p>]]>
        </description>
    </item>
    <item>
        <title>(2026) Habib Geryes | Pediatric PCCT protocols and radiation exposure</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/490/2026-habib-geryes-pediatric-pcct-protocols-and-radiation-exposure</link>
        <pubDate>Fri, 28 Aug 2026 06:54:37 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">490@/de/discussions</guid>
        <description><![CDATA[<div><div><p>This multicenter retrospective study evaluated 6,464 PCCT examinations in 4,772 pediatric patients across three French institutions and provides practical acquisition/reconstruction protocols across ten body regions.</p><p><br />
The cohort had a mean age of 7.9 ± 5.5 years (4 days–17.9 years); nearly one-third of examinations involved children &lt;15 kg and &lt;4 years. Chest and head examinations were the most frequent.</p><p><br />
Protocols were explicitly designed to balance radiation-dose reduction, diagnostic image quality, motion reduction, and workflow efficiency, with dose analysis stratified into five weight groups.</p><p><br />
The authors conclude that the results support safe clinical integration of pediatric PCCT, while emphasizing that the proposed protocols are a flexible framework for local optimization rather than fixed universal settings.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fwww.sciencedirect.com%2Fscience%2Farticle%2Fpii%2FS2211568425002153%3Fvia%253Dihub" rel="nofollow noopener ugc" role="button">Access publication</a></p><h2 data-id="protocols">Protocols</h2><p>We created an cheat sheet from the protocol table via AI:</p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/87N60D9P975B\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:1872062,&quot;width&quot;:1054,&quot;height&quot;:1492,&quot;displaySize&quot;:&quot;medium&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F87N60D9P975B%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1662,&quot;dateInserted&quot;:&quot;2026-08-28T07:21:42+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2F87N60D9P975B%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/87N60D9P975B/image.png" alt="image.png" height="1492" width="1054" data-display-size="medium" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/87N60D9P975B/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/87N60D9P975B/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/87N60D9P975B/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/87N60D9P975B/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/87N60D9P975B/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/87N60D9P975B/image.png 2000w, https://us.v-cdn.net/6038601/uploads/87N60D9P975B/image.png" sizes="100vw" /></a>
    </span>
</span>
<p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Result</strong></p></th></tr><tr><td><p><strong>Patients</strong></p></td><td><p><strong>4,772</strong></p></td></tr><tr><td><p><strong>PCCT examinations</strong></p></td><td><p><strong>6,464</strong></p></td></tr><tr><td><p><strong>Mean age</strong></p></td><td><p><strong>7.9 ± 5.5 years</strong></p></td></tr><tr><td><p><strong>Age range</strong></p></td><td><p><strong>4 days–17.9 years</strong></p></td></tr><tr><td><p><strong>Mean weight</strong></p></td><td><p><strong>29.6 ± 20.3 kg</strong></p></td></tr><tr><td><p><strong>Weight range</strong></p></td><td><p><strong>2–79 kg</strong></p></td></tr><tr><td><p><strong>Children &lt;15 kg / &lt;4 years</strong></p></td><td><p><strong>2,110 examinations (32.64%)</strong></p></td></tr><tr><td><p><strong>Most frequent examination</strong></p></td><td><p>Chest: <strong>2,168 (33.54%)</strong></p></td></tr><tr><td><p><strong>Second most frequent</strong></p></td><td><p>Head: <strong>1,675 (25.91%)</strong></p></td></tr><tr><td><p><strong>Noncontrast head CTDIvol</strong></p></td><td><p><strong>16.1–24.0 mGy</strong> across weight groups</p></td></tr><tr><td><p><strong>Noncontrast head DLP</strong></p></td><td><p><strong>311–608 mGy·cm</strong></p></td></tr><tr><td><p><strong>Noncontrast chest CTDIvol</strong></p></td><td><p><strong>0.3–0.8 mGy</strong></p></td></tr><tr><td><p><strong>Noncontrast chest DLP</strong></p></td><td><p><strong>7–27 mGy·cm</strong></p></td></tr><tr><td><p><strong>Temporal bone CTDIvol</strong></p></td><td><p><strong>22.7–33.8 mGy</strong></p></td></tr><tr><td><p><strong>Dose relationship</strong></p></td><td><p>CTDIvol and DLP increased with patient weight</p></td></tr></table></div><p>The patient distribution and examination frequencies are reported in the study results.  Dose levels varied substantially by examination and patient weight; temporal-bone examinations had particularly high CTDIvol because of UHR/high-kV acquisition, whereas noncontrast chest imaging operated at very low dose levels. </p><p>For example, pooled across weight groups, <strong>noncontrast chest PCCT</strong> had a median CTDIvol of <strong>0.4 mGy</strong> and DLP of <strong>11.9 mGy·cm</strong>, while the dedicated <strong>low-dose chest</strong> protocol reached approximately <strong>0.1 mGy</strong> and <strong>4.0 mGy·cm</strong>, respectively. </p><p></p><h3 data-id="practical-take-home">Practical take-home</h3><p>This is primarily a <strong>protocol and dose-reference paper</strong>, rather than an image-quality superiority study. Its major value is the large real-world pediatric dataset and the detailed, indication-specific protocol framework. PCCT allows aggressive pediatric dose optimization through <strong>weight-/task-adapted CARE keV, Sn filtration, fast-pitch acquisition, UHR where spatial resolution is necessary, and spectral reconstructions from contrast-enhanced QuantumPlus scans</strong>.  </p><p><strong>Source:</strong> Habib Geryes B, Moalla S, De Saint Etienne E, et al. <em>Photon-counting CT in pediatric patients: A multi-institutional evaluation of protocols and radiation exposure in 4,772 patients.</em> <strong>Diagnostic and Interventional Imaging.</strong> 2026;107(5):186–197. doi:10.1016/j.diii.2025.11.010.</p>]]>
        </description>
    </item>
    <item>
        <title>(2022) Mergen | Virtual noncontrast abdominal imaging with PCCT</title>
        <link>https://connect.siemens-healthineers.com/de/discussion/487/2022-mergen-virtual-noncontrast-abdominal-imaging-with-pcct</link>
        <pubDate>Thu, 27 Aug 2026 17:36:07 +0000</pubDate>
        <category>Publication protocols</category>
        <dc:creator>Daniel Stromer</dc:creator>
        <guid isPermaLink="false">487@/de/discussions</guid>
        <description><![CDATA[<div><div><p>Virtual noncontrast (VNC) images reconstructed from contrast-enhanced PCCT provided accurate abdominal attenuation values compared with true noncontrast (TNC) imaging, supporting the potential to omit a separate TNC acquisition.</p><p><br />
Across all measurements, VNC attenuation errors were &lt;5 HU in 76% and &lt;10 HU in 95%; there was no meaningful overall difference between VNC reconstructed from arterial versus portal venous acquisitions.</p><p><br />
VNC had inferior image quality and higher/blotchier noise than TNC, but remained diagnostically usable in 99–100% of patients.</p></div></div><p><a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fpubs.rsna.org%2Fdoi%2F10.1148%2Fradiol.213260" rel="nofollow noopener ugc" role="button">Access publication</a></p><p></p><span data-embedjson="{&quot;url&quot;:&quot;https:\/\/us.v-cdn.net\/6038601\/uploads\/FZEKUMKR08AW\/image.png&quot;,&quot;name&quot;:&quot;image.png&quot;,&quot;type&quot;:&quot;image\/png&quot;,&quot;size&quot;:1376583,&quot;width&quot;:2000,&quot;height&quot;:1150,&quot;displaySize&quot;:&quot;large&quot;,&quot;float&quot;:&quot;none&quot;,&quot;downloadUrl&quot;:&quot;https:\/\/connect.siemens-healthineers.com\/api\/v2\/media\/download-by-url?url=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FFZEKUMKR08AW%2Fimage.png&quot;,&quot;active&quot;:true,&quot;mediaID&quot;:1657,&quot;dateInserted&quot;:&quot;2026-08-27T17:35:43+00:00&quot;,&quot;insertUserID&quot;:14,&quot;foreignType&quot;:null,&quot;foreignID&quot;:null,&quot;embedType&quot;:&quot;image&quot;,&quot;embedStyle&quot;:&quot;rich_embed_card&quot;}">
    <span>
        <a href="https://connect.siemens-healthineers.com/de/home/leaving?allowTrusted=1&amp;target=https%3A%2F%2Fus.v-cdn.net%2F6038601%2Fuploads%2FFZEKUMKR08AW%2Fimage.png" rel="nofollow noopener ugc" target="_blank">
            <img src="https://us.v-cdn.net/6038601/uploads/FZEKUMKR08AW/image.png" alt="image.png" height="1150" width="2000" data-display-size="large" data-float="none" data-type="image/png" data-embed-type="image" srcset="https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=300, width=300/6038601/uploads/FZEKUMKR08AW/image.png 300w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=600, width=600/6038601/uploads/FZEKUMKR08AW/image.png 600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=800, width=800/6038601/uploads/FZEKUMKR08AW/image.png 800w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1200, width=1200/6038601/uploads/FZEKUMKR08AW/image.png 1200w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=1600, width=1600/6038601/uploads/FZEKUMKR08AW/image.png 1600w, https://us.v-cdn.net/cdn-cgi/image/quality=80, format=auto, fit=scale-down, height=2000, width=2000/6038601/uploads/FZEKUMKR08AW/image.png 2000w, https://us.v-cdn.net/6038601/uploads/FZEKUMKR08AW/image.png" sizes="100vw" /></a>
    </span>
</span>
<p>Figure 2: CT attenuation measurements in a 58-year-old man referred for follow-up examination after an acute type B aortic dissection. Axial CT images acquired with photon-counting detector CT depict exemplary region-of-interest (ROI) measurements on venous virtual noncontrast reconstructions. avg = average attenuation within region of interest; d = diameter of region of interest.</p><p></p><h2 data-id="protocol">Protocol</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Detail</strong></p></th><th><p><strong>Value</strong></p></th></tr><tr><td><p><strong>Scanner</strong></p></td><td><p>NAEOTOM Alpha</p></td></tr><tr><td><p><strong>Detector</strong></p></td><td><p>Dual-source, two cadmium telluride PCDs</p></td></tr><tr><td><p><strong>Scan area</strong></p></td><td><p>Thoracoabdominal</p></td></tr><tr><td><p><strong>Scan phases</strong></p></td><td><p>TNC + arterial + portal venous</p></td></tr><tr><td><p><strong>Acquisition mode</strong></p></td><td><p>Single-source multienergy, <strong>QuantumPlus</strong></p></td></tr><tr><td><p><strong>Tube voltage [kV]</strong></p></td><td><p>120</p></td></tr><tr><td><p><strong>Tube current</strong></p></td><td><p>Automated modulation, CARE Dose4D / CARE kV</p></td></tr><tr><td><p><strong>IQ level</strong></p></td><td><p>68</p></td></tr><tr><td><p><strong>Rotation time [s]</strong></p></td><td><p>0.25</p></td></tr><tr><td><p><strong>Pitch</strong></p></td><td><p>1.2</p></td></tr><tr><td><p><strong>Detector slice collimation</strong></p></td><td><p>144 × 0.4 mm</p></td></tr><tr><td><p><strong>Contrast</strong></p></td><td><p>70 mL iopromide, 370 mg I/mL</p></td></tr><tr><td><p><strong>Injection rate</strong></p></td><td><p>4 mL/s</p></td></tr><tr><td><p><strong>Arterial timing</strong></p></td><td><p>Bolus tracking; 140 HU threshold at 90 kV + 12-s delay</p></td></tr><tr><td><p><strong>Portal venous timing</strong></p></td><td><p>70 s after contrast injection</p></td></tr><tr><td><p><strong>Reconstruction kernel</strong></p></td><td><p>Br36</p></td></tr><tr><td><p><strong>Iterative reconstruction</strong></p></td><td><p>QIR 4</p></td></tr><tr><td><p><strong>Slice thickness [mm]</strong></p></td><td><p>2.0</p></td></tr><tr><td><p><strong>Reconstruction increment [mm]</strong></p></td><td><p>1.6</p></td></tr><tr><td><p><strong>Reconstruction plane</strong></p></td><td><p>Axial</p></td></tr><tr><td><p><strong>VNC source phases</strong></p></td><td><p>Arterial and portal venous</p></td></tr><tr><td><p><strong>Reference standard</strong></p></td><td><p>True noncontrast (TNC) images</p></td></tr></table></div><p></p><h2 data-id="key-results">Key Results</h2><div><table><colgroup><col /><col /><col /></colgroup><tr><th><p><strong>Parameter</strong></p></th><th><p><strong>Main finding</strong></p></th></tr><tr><td><p><strong>Study population</strong></p></td><td><p>100 patients; mean age 72 ± 10 years; 81 men</p></td></tr><tr><td><p><strong>Overall VNC error &lt;5 HU</strong></p></td><td><p><strong>76%</strong> of measurements</p></td></tr><tr><td><p><strong>Overall VNC error &lt;10 HU</strong></p></td><td><p><strong>95%</strong> of measurements</p></td></tr><tr><td><p><strong>Arterial vs venous VNC error</strong></p></td><td><p>3.3 vs 3.5 HU; <strong>P = .16</strong></p></td></tr><tr><td><p><strong>Left liver attenuation error</strong></p></td><td><p><strong>1.4 ± 1.3 HU arterial VNC; 1.5 ± 1.2 HU venous VNC</strong></p></td></tr><tr><td><p><strong>Liver error &lt;5 HU</strong></p></td><td><p>Left lobe <strong>99%</strong>; right lobe <strong>98%</strong> for both arterial and venous VNC</p></td></tr><tr><td><p><strong>Diagnostic subjective quality</strong></p></td><td><p><strong>99–100%</strong></p></td></tr><tr><td><p><strong>VNC noise</strong></p></td><td><p><strong>33% higher</strong> than TNC in phantom</p></td></tr><tr><td><p><strong>Noise texture</strong></p></td><td><p>VNC more blotchy; average spatial frequency <strong>0.14 vs 0.20 mm⁻¹</strong></p></td></tr><tr><td><p><strong>Spatial resolution</strong></p></td><td><p>Similar overall; TTF50 <strong>0.29 mm⁻¹</strong> for both</p></td></tr><tr><td><p><strong>Detectability index</strong></p></td><td><p>VNC approximately <strong>15% lower</strong>, but always <strong>&gt;20</strong></p></td></tr></table></div><p>The liver showed particularly small VNC attenuation errors, with left-lobe mean errors around <strong>1.5 HU</strong>.  Phantom testing confirmed higher and blotchier VNC noise but similar middle-frequency spatial resolution and a detectability index consistently above 20. </p><p><strong>Take-home:</strong> PCCT VNC can provide quantitatively accurate abdominal attenuation without a dedicated noncontrast acquisition. The principal trade-off is higher, blotchier image noise and somewhat poorer subjective image quality, although diagnostic quality remained very high. </p><p><strong>Source:</strong> Mergen V, Racine D, Jungblut L, et al. <em>Virtual Noncontrast Abdominal Imaging with Photon-counting Detector CT.</em> <strong>Radiology.</strong> 2022;305(1):107–115. doi:10.1148/radiol.213260.</p>]]>
        </description>
    </item>
   </channel>
</rss>